RE: APC/PE

From: Kevin Holmes (KHOLMES@atlas.niaid.nih.gov)
Date: Tue Dec 02 1997 - 12:18:35 EST


Gene Pizzo wrote:
	Furthermore, why on the FACSCalibur is it necessary to align the
two beams
	in time when on the FACStar Plus the electronic separation of
the two beams
	in time is considered a necessity?
The difference between the Calibur and the FACStar Plus is that on the
Calibur (with a 2nd laser) you are trying to detect two dyes that, if
interlaser compensation is not done, will both generate a signal in the
same (FL4) PMT.  Specifically, you are trying to use PE-Cy5 and APC
simultaneously; the Cy5 component of the PE-Cy5 dye will be excited by
the 633nm laser and generate a signal in FL4.  What you are doing is
actually adjusting the time that the signal from the 1st laser (PE-Cy5)
is delayed in order to subtract it (compensate) from the signal off of
the 2nd laser (Cy5 component of PE-Cy5).  You are not, as I think you
are suggesting, aligning the two beams to be colinear; they are still
separated in time.

Kevin L. Holmes, Ph.D.
Head, Flow cytometry Unit
Office of the Scientific Director
Bldg. 7, Room 01
NIAID, NIH

Phone: 301-496-9071
FAX:  301-402-4532
	Email: kholmes@atlas.niaid.nih.gov

	Colleagues,

	I'll ask the question again...it hasn't been answered here yet,
twice.
	It wasn't answered in BD class when I asked it.
	It wasn't answered by the service engineers I've asked.
	Maybe no one has an answer sort of like that question in
computer class
	years ago "does 'bit' derive from the first two letters of
binary or the 
	last
	two letters of digit?"

	We all understand that the Cy5 of the PE-Cy5 tandem can be
excited
	by the red laser, that's the easy part. But if the signal
generated from
	the two beams is separated in both space and time why should
there
	be a compensation necessity?

	Furthermore, why on the FACSCalibur is it necessary to align the
two beams
	in time when on the FACStar Plus the electronic separation of
the two beams
	in time is considered a necessity?

	Gene Pizzo/UCONN Health 