Litron@aol.com wrote: > > I have been having an on again-off again problem that maybe somebody can help > me with. When I push the on button on the control module to start the laser, > it immediately shuts off. I can do this 5, 10, sometimes 30 times before it > finally stays on. H20 pressure is OK, no indicator lights come on, our > Coherent tech-who is great-is stumped. It's been suggested that we may need > a new control module-cost $2500. Does anybody have one in good condition > they are interested in selling, or does anybody have any other suggestions > for me? All info will be much appreciated. > > Dorothea Torous > litron@aol.com Dorothea, Not only the H2O pressure is important, but also the H2O flow in the cooling system. If the H2O flow is close to the required minimum the laser will act like you have described. Check your cooling system on obstacles that can down regulate the H2O flow and keeping the H2O pressure on a "normal" level. We have seen the same problem in our cooling system after inserting two other lasers for a new flowcytometer. We could keep the H2O pressure on a "normal" level, but we could not start the third laser for the reason described here. Maarten van der Keur Department of Hematology University Hospital Leiden