From: John Ladasky (ladasky@leland.Stanford.EDU)
Date: Mon Jun 16 1997 - 17:29:13 EST


>Date: Mon, 16 Jun 1997 08:17:33 -0400 (EDT)
>From: Matthew J Shaw <shawmj@battelle.org>
>Subject: Viscosity Effects on FCM Analysis
>
>     All,
>     
>     Does anyone have a good feel for the effects of higher viscosity 
>     fluids being analyzed via flow cytometric means?  For example, the 
>     viscosity of water is 1 centipoise, and I assume that cells that are 
>     within water can be reliably analyzed.  I believe that the viscosity 
>     of blood is perhaps 3 centipoise; again, I assume that cells within 
>     this matrix can be reliably analyzed.  However, if the matrix is say 
>     10 centipoise, there may be a problem with forcing the matrix through 
>     the FCM nozzle.  Maybe a problem doesn't exist until the matrix is 100 
>     centipoise?  Has anyone done a study like this on any flow cytometer, 
>     or have a good feel for this effect?  What is the upper viscosity 
>     limit before any problems may occur?  (I realize that one could dilute 
>     the sample with a low-viscosity liquid until the desired viscosity is 
>     reached - I'd rather not have to do that!) I would appreciate any help 
>     on this matter.
>     
>     Matt Shaw

	Many years ago, I performed an experiment with a FACScan that sup-
ports Matt's hypothesis that viscous fluids may not flow as readily through
the flow cytometer.  My "viscous solution" was detergent-solubilized nuclei
from the Jurkat cell line, stained with propidium iodide.  The stock solution
was fairly concentrated, as I recall -- perhaps 10 million cells/ml.  I did
not have the means for measuring the viscosity of this solution.  I tried
this at full strength and at various dilutions.  I weighed the tubes on an
analytical balance, ran them on the FACScan for five minutes, and weighed
them again.  I used the low flow rate, which pumps 12 microliters/minute 
when the sample tube contains just plain water.  Over the course of the
five minutes, I watched the acquisition event rate in order to be sure that
any drop that I might see in the *mass* flow rate (microliters/minute) was
not the result of temporary clogs in the sample injection tube.  I think
that I did get such clogs at the higher cell concentrations.  However, even at
the lower concentrations, there was clearly a drop in the mass of liquid 
transferred, even when there were no hiccups detectable in the event rate.

	Hope that helps!

							- John Ladasky