Hi Ryan... We've had the same problem as you when attempting to analyze apoptosis in human cells using ethanol fixation and staining with PI or other DNA binding agents, namely the accumulation of an apoptotic peak with a very low fluorescence intensity. It is usually impossible to distinguish apoptotic cells from debris using a single parameter linear scale; log scaling is often only somewhat successful for distinguishing apoptotic cells from debris. It appears that the amount of DNA extruded by apoptotic cells is quite variable from cell type to cell type; this appears to be particularly true for human cells, and is undoubtedly related to the type of DNA damage experienced during apoptotic death. We generally limit our use this technique in mouse lymphoid tissues, since they give a nice, clean apoptotic peak at an index of about 0.5 of the diploid G1 peak. If you really want to use single DNA dye staining to look at apoptosis, analyzing on a log scale for fluorescence with forward scatter as a second parameter is probably the best way to go. Apoptotic cells will show somewhat reduced FSC signal but will be distinguishable from debris (which should be quite small). Also be aware that some apoptotic stimuli can cause extensive cell blebbing, which will reduce your apparent dead cell yield (since these blebs will likely appear as debris). Most importantly, make sure that the population you think is apoptotic actually is - if dying cells constitute the majority of objects in your culture, simple examination under a fluorescence or confocal microscope (or EM) may be sufficient. If it is a smaller number, it would probably be best to sort the cells and then check by microscopy (I frankly think everyone who does any sort of flow assay for apoptosis should do this at least once as an essential control). You should also probably do another type of cell death assay in parallel (at least a couple of times) to confirm the presence of cell death (such as two-dye exclusion, TUNEL, FITC-annexin, etc.). If you have continued difficulties with single-dye methods, it might be best to switch - they are being used less as of late since they are not as good at distinguishing dead cells from debris in some cell types, as well as the variable position/appearance of the dead cell region. Also check Dr. Darzynkiewicz's protocol for single-dye staining using adjusted phosphate-citrate buffers prior to staining - this technique attempts to control the amount of DNA extrusion and thus the position of the apoptotic peak. Hope this helps. If you want, FAX me a copy of your data and I can compare it to what we've seen for apoptotic cells. Cheers! Bill Telford ______________________________ Hospital for Special Surgery/ Cornell University Medical College Flow Cytometry Core Facility New York, NY 10021 phone: (212) 606-1925 FAX: (212) 717-1192 On Mon, 9 Jun 1997, Ryan Hung wrote: > > I've just started using FACS analysis for analyzing apoptosis in PBMC a > couple of months ago. I figured that I would first try the inexpensive > technique of PI staining. However, to date, having tried both direct > staining with PI (with sodium citrate, Triton X-100, RNAase A) and > ethanol-fixation prior to PI staining, I have not been able to get either > to work reliably. > > My model system itself consists of PBMC extracted from healthy human > donors at a concentration of 3x10^6 cells/ml in 1ml of RPMI growth medium > in 24 well plates. I'm using camptothecin in a concentration of 5 ug/ml > as my positive control of apoptosis. I incubate for between 14-18h at > 37oC in 5% CO2. > > With direct staining with PI, I've typically obtained two closely-spaced > high fluorescent intensity peaks when reading FL3 on a histogram plot. I > see virtually no lower fluorescent intensity peaks. > > With ethanol fixation, I've had problems with gating when using a linear > scale on FSC, so I've tried using a log scale. I understand from previous > discussion that this may be incorporating significant amounts of cellular > debris. What I observe in terms of histograms are a single > high-fluorescence peak, with a lower fluorescent peak at about 1/5 the > intensity of the high peak, rather than the expected 1/2. > > Since my results from both techniques don't seem to correspond well with > what I've found in literature, I've been finding it difficult to analyze. > Any ideas or suggestions would be much appreciated. > > Ryan. > > _/ \__/ \__/ \__/ \__/ \__/ \__/ \__/rhung@vcn.bc.ca__/ \__/ \__/ > \_Apoptosis=programmed cell death/ \__/ \rwhung@netinfo.ubc.ca \__/ \__ > _/ --you can't live without it!/ \__/ \http://www.vcn.bc.ca/people/rhung > \__/ \__/ \__/ \__/ \__/ \__/ \__/ \My words Copyright (C) 1997 \__ > >
This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:49:50 EST