FACScan cleaning and maintenance

From: Andreas.Christaras (christar@uni-duesseldorf.de)
Date: Wed Apr 16 1997 - 18:09:43 EST


Dear Flow-Folk,

I have some questions concerning cleaning and maintaining a BD FACScan.
Unfortunately, changing the technicians maintaining and sometimes repairing
the FACScan obviously implies changes in cleaning and maintenance
recommendations and regulations. Initially, following procedure for
cleaning the hydraulic system and measurement capillary on a regular basis
was recommended:

A falcon tube with 0.1% solution of sodium hypochlorite was prepared and
then run over 10-15 minutes with flow control system set to "high flow" and
function switch set to "measurement". Subsequently a tube filled with
destilled water was run over 10-15 minutes (same settings as sodium
hypochlorite solution). This recommendation was exactly the one I found in
the German version of FACScan user's manual by BD.
Later, our lab was recommended to replace the sodium hypochlorite solution
with 0,1% 7 flow or 7X solution and to follow the time limits and settings
as mentioned before for sodium hypochlorite based cleaning.
Last year, after several incidents of hydraulic system block by cell
detritus, recommendation for cleaning was again changed to 0.1% sodium
hypochlorite solution (no change on instrument settings).
Now, as happened yesterday, cleaning recommendation was changed again to
7-flow or 7-X in a 0.1% solution.

If any of you have some idea, information, or experiences indicating the
right and correct (and even the best functioning) method of cleaning the
FACScan, please drop a note. Any incoming suggestions or comments are
appreciated.

Thanks in advance,
Andreas

<><><><><><><><><><><><><><><><><><><><><><KKK   KKK MMM          MMMM TTTTTTT
Andreas Christaras                         KKK  KKK  MMMM        MMMMM TTTTTTT
Heinrich-Heine University Duesseldorf      KKK KKK   MMMMM      MMMMMM   TTT
Children's Hospital                        KKKKKK    MMM MMM   MMM MMM   TTT
Dept of pediatric hematology & oncology    KKKKK     MMM  MMM MMM  MMM   TTT
Lab of experimental hematology (KMT-Labor) KKKK      MMM   MMMMM   MMM   TTT
Geb. 14.82 (Bldg.)                         KKKKK     MMM    MMM    MMM   TTT
Moorenstrasse 5 - 40225 Duesseldorf        KKK KKK   MMM     M     MMM   TTT
GERMANY                                    KKK  KKK  MMM           MMM   TTT
                                           KKK   KKK MMM           MMM   TTT
FON  : +49-211-81-16103
         
FAX  : +49-211-81-16206 or -16191          BLOOD  IS  OUR  O B S E S S I O N 
EMAIL: christar@uni-duesseldorf.de                                            
<><><><><><><><><><><><><><><><><><><><><><><><><><><><><><><><><><><><><><><>



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:49:40 EST