Re: Bacteria

From: Hazel Davey (hlr@aber.ac.uk)
Date: Wed Apr 16 1997 - 08:16:53 EST


On 14 Apr 97 at 9:28, Peter Lopez wrote:

> 
> Hello,
> 
>  I have been doing some E.Coli analysis and sorting and was
> wondering if anyone had a favorite sheath fluid to use for these
> bugs. I am looking to make them as happy as possible during sort
> cloning. Thanks for any tips.

Many of the commercial preps are unsuitable for bacteria because of
the high sodium content - they are designed to keep animal (blood)
cells happy. I haven't done sorting with coli but Micrococcus luteus
seem ok (i.e. grow afterwards) when sorted in a sheath fluid
containing 150mM KCl and 10mM Hepes (pH 6.8 with KOH). This was with 
cells that weren't exactly happy to begin with :

Kaprelyants AS, Mukamolova GV, Davey HM, and Kell DB: Quantitative
analysis of the physiological heterogeneity within starved cultures of
Micrococcus luteus using flow cytometry and cell sorting. Appl. Env.
Microbiol. 62:1311-1316, 1996.


Haze
----------------------------------------------------------