RE: flow assay for hereditary spherocyto

From: Leary, James (jleary@mspo3.med.utmb.edu)
Date: Wed Mar 26 1997 - 09:26:00 EST


I would not expect light scatter pulse width to work.  However, a   
combination of forward scatter, side scatter and pulse rise-time   
time-of-flight should work.  Pulse rise-time is sensitive to the shape,   
not just the diameter, of cells.  If you have access to this measurement   
on your flow cytometer, use start and stop thresholds of 5% and 35%   
respectively for greatest sensitivity.

 -----Original Message-----
From: Yashpal Agrawal [SMTP:yagrawal@messi.uku.fi]
Sent: Tuesday, March 25, 1997 5:43 PM
To: cyto-inbox
Subject: flow assay for hereditary spherocytosis


I tried to measure pulse width on a FACStar machine for this purpose. The   
method was not able to differentiate from normal RBC's.
Hope this is useful.
Yash Pal Agrawal
Kuopio, Finland