A week ago I have asked about solving my troubles with the laser in FACSort. Since BD technician recommended the laser replacement (in spite of the laser current still was not so high-around 7 Amps) I asked you about your opinion. Many of you suggested insufficient laser (cytometer) cooling: (thermally unstable Status card; a component in the power supply is overheating and cutting out; if a component is failing due to heat, it is likely to be some small electronic bit rather than the laser tube). There are no obstructions to the vent in the back of the instrument and fans are working properly. I decided to clean the whole instrument from dust using compressed air. The laser inside and the filter on its side was full of dust!!! After that cleaning procedure and 1 week running the cytometer I have not observed the failure again. Thank you, Regards, -- _______________________________________________________________ Karel Drbal Laboratory of Leukocyte Antigens Institute of Molecular Genetics Academy of Sciences of the Czech Republic Videnska 1083 142 20 PRAHA 4 Czech Republic, Europe voice: +42-2-4752589 fax: +42-2-4727979 e-mail: drbal@biomed.cas.cz WWW: http://leuko.biomed.cas.cz