Several items to consider: 1) If you are trying to measure the change in the average population intensity, then using the ratio of the medians is a valid way to go. 2) Logamps aren't perfect, but the modern ones aren't "that" bad. We have done extensive measurements on our BD machines (Facscan, Facstar+, Vantage) and find that the error in their logamps is no greater than +-5%, except right at the lower and upper edges (the same is true for the MoFlo - we have not looked at a Coulter machine in years, but would be surprised if they are much different). So in a ratio, the error shouldn't be greater than +-10%. I am confused about your relative fluoescence scale. A four decade log amp should go from 1-10K (or .1-1K, etc.), not 0-10K. This distortion of the low decade could influence the results. 3) In order to compare measurements from day-to-day on the same machine, standardization is important. We standardize by setting a "standard particle" (i.e. bead) to the same position on each channel for every experiment. This requires slightly different pmt voltages. The relative sensitivity = 1/gain, where gain = (V/Vo)**7.4, where Vo is the standard pmt voltage and V is the voltage used. This is an accurate measure of how well the instrument is aligned overall. A 20% difference can indicate setup problems. 4) It is important to know the inherent error in the biological system. Do several preparations of the same sample measured at the same time give the same results? -Marty Bigos Stanford Shared FACS Facility