We have stained platelets many different ways. Adding prediluted antibody to 1:100 diluted PRP (platelet rich plasma, prepared by a gentle spin to settle out RBC & leuks) has been the most satisfactory to us. Some people like to fix the cells, but we have noticed changes in some glycoprotein expression, and only fix when absolutely necessary. A fellow lab simply dilutes 5 - 8 ul whole blood in buffer and adds stock Ab, inc. 30 min., and runs on cytometer. A few other hints (for the new to platelets): washing and lysing protocols can sometimes activate the platelets, as you may know, so we try to avoid this. Not washing away excess Ab can lead to higher nonspecific fluorescence, so antibodies should be titered to optimum concentration per platelet number in your lab prior to establishing a routine protocol. Always add negative control Ab at the same ug amount as positive Ab, and verify the same isotype and molar ratio of dye to protein. BD makes a good PE platelet neg CTL (cat # 340013) that has reduced non-bound PE in the suspension, but we have used others that seem to work as well. Also, try to avoid using EDTA as the anticoagulant. GPIIbIIIa expression is Ca dependent to some degree, as well as some activation markers. Na3Citrate is best. Miller and Bode have shown that triggering on linear side scatter (visualizing on log forward scatter) gives the best platelet distribution. I'm sure there are other opinions in the field, but the above has worked well for us. I hope you find it useful.