Chris Hathaway asks:

"....would like to evaluate circulating neutrophils, specifically 
quantify them according to their activation state.  Is there any way 
to determine the number (percentage) of cells that are primed, 
activated, etc. from a blood sample? "

Chris, this is a really easy thing to do in flow cytometry - there 
are several probes that are excellent for the job, namely 
dihydrorhodamine 123, hydroethidine and dichlorofluorescin diacetate. 
These probes allow you to measure the respiratory burst of 
neutrophils. The methods are very straight forward. You incubate hte 
cells with the probe, stimulate the cells with an appropriate 
stimulant (PMA, fmlp, etc) and measure the change in fluorescence 
after a few minutes. The resting cells will exhibit a fluorescence 
level equal to the basal respiration of the cells. If your cells are 
activated, then this "resting" fluorescence will be high. We observed 
this some years ago when measuring hte difference between "resting" 
ie unstimulated neutrophils taken from gum diseased periodontal sites 
as opposed to "clean" sites. (Infect.Immun.56:156-160 (1988))

Essentially you can measure several reactive oxygen or reactive 
nitrogen molecules using fluorogenic probes. Detailed methods can be 
found in the Handbook of Flow Cytometry Methods (Wiley Liss), or 
Methods in Cell Biology, 41:437-447, 1994.

If you have specific questions give me a call.