This material was originally published in the Purdue Cytometry CD-ROM Series,volume 4
Modification
of antigen expression in apoptotic cells: a single laser methodology.
Filippo Renò and Stefano
Papa
Centro di Citometria, Istituto
di Scienze Morfologiche, Università di Urbino, Italy
Tel /Fax: +39/722/304242 - E-mail:
reno@bio.uniurb.it papa@bio.uniurb.it.
1. Introduction
Many cytometric methods have
been used in recent years to identify cells in different stages of apoptosis
(early, middle, and late) (4,7). We describe a new, easy method which allows
the simultaneous identification of apoptotic sub-populations and the presence/absence
of surface antigens on their plasmatic membrane during apoptotic development.
In this protocol we use fluorescein diacetate (FDA) (10) or Annexin V-Fitc
(8) to discriminate apoptotic cells and Trypan Blue to identify necrotic
cells. Trypan Blue, normally used for the optical detection of dead cells,
is also a fluorescent dye emitting a red fluorescence (l>
620 nm) when excited by the blue light of an argon laser tuned on
488 nm wavelength (11) .
2. Protocol
2.1 Materials
Human myeloid cell line HL-60
RPMI 1640 (A1)
Phosphate buffer (PBS) pH 7.4 (A1)
Hepes buffer, pH 7.4 (A1)
Fluorescein diacetate (FDA) (A1,A2)
Trypan Blue solution (0.4% v/v)
(A1,A2)
Camptothecin (A2)
Annexin V-FITC (A2)
Anti-CD11a PE (A2)
Centrifuge (A3)
Flow cytometer (argon ion laser)
(A3)
2.2 Methodology
Basic Protocol
Note: Below, we describe two
protocols in which we utilize HL-60 human tumoral cells. These protocols
are also adaptable to different cellular lines (Molt-4, Raji, Jurkat, Daudi
and K562), human lymphocytes and different apoptotic triggers (UVB, Staurosporine,
hyperthermia, etc.).
1. Induce apoptosis in HL-60
cells in logarithmic growth phase. Incubate cells (1 x 105 cells/ml)
with 0.15 µM camptothecin, a topoisomerase I inhibitor, for 6 hours
at 37°C in a CO2 incubator.
2. Collect cells in 15 ml centrifuge
tubes and wash twice with PBS. Spin at 1500 rpm (300 x g) for 5 minutes.
3. Resuspend cells at 1 x 106
cells/ml concentration in PBS.
4. Incubate 2.5 x 105
cells (250 µl) with 5 µl anti-CD11a antibody labelled with
PE for 20 minutes at room temperature.
5. Wash twice with PBS and check
the positivity of the control population and apoptotic population for LFA-1
antigen.
6. Add 5 µl of 1µg/ml
FDA solution prepared in PBS shortly before use, to cells stained with
anti-CD11a PE.
7. Gently resuspend cells and incubate
at room temperature for 5 minutes.
8. Add 25 µl Trypan Blue at
a final concentration of 800 µM. Incubate for 5 minutes and then
quantify the apoptotic and dead cells by flow cytometry.
Alternative Protocol
1. Follow steps 1 to 5 of the Basic
Protocol replacing PBS buffer with Hepes buffer in every step.
2. Add 10 µl of Annexin V-
Fitc at working dilution (1 µg/µl) to the stained cells.
3. Incubate for 15 minutes on the
bench.
4. Analyze by flow cytometry and
compensate the Annexin V green fluorescence with the anti-CD11a orange
fluorescence.
5. Add 25 µl Trypan Blue at
a final concentration of 800 µM. Incubate for 5 minutes and then
quantify the apoptotic and dead cells by flow cytometry.
3. Commentary
3.1 Background Information
Decreases in some cell surface antigens
have been seen in apoptotic cells. This occurs in human lymphocytes with
the CD45 antigen (1) in neutrophils for different antigens (5,6,8) and
in thymocytes for the CD4 and CD8 antigens (2,13). The reason behind the
shedding of antigens during apoptosis is still not clear and could perhaps
be correlated with a change in the plasmatic membrane structure. Nevertheless
it is possible to postulate a role for this surface antigen in some apoptotic
processes or in their inhibition (3,15). Many attempts have been made to
find a protocol capable of discriminating the apoptotic stage and the phenotype
in homogeneous and heterogeneous cellular populations using dual laser
cytometer (2,12,14) or fixed cells (9). Instead, our method works using
a single laser cytometer, easy to use and to find in the majority of clinical
and research cytometry facilities. Moreover, we always use unfixed cells
thus avoiding all problems related to antigen loss or accessibility after
the fixation process. Another advantage of using unfixed cells is the possibility
to follow the development of apoptosis also using other parameters (e.g.
scatter signals) (4) and to check, step by step, the presence of certain
antigens on apoptotic subpopulations during the whole apoptotic process.
Finally, it is also possible to sort the apoptotic cells and analyze their
morphology using microscopy (optical or electronic).
3.2 Critical Parameters
This method should not be used if
you need fetal calf serum proteins or albumin in your medium. Trypan Blue
has a great affinity for proteins and the discrimination between live and
dead cells becomes less evident if a large amount of extracellular proteins
is present in the assay microenvironment (11). It also important to avoid
the exposure of cells to Trypan Blue for a period longer than 30 minutes.
In this case it is possible to observe an increase in the dead cell population
(Trypan Blue positive) due to the Trypan toxicity.
3.3 Troubleshooting
The FDA staining can sometimes result
in a weak fluorescence. This can be due either to an error in the preparation
of the FDA staining solution, which should be prepared shortly before use
every time the experiment is carried out, or to an insufficient quantity
of the non-specific esterase substrate. For this reason, we suggest performing
preliminary experiments in order to find the most efficient FDA concentration
for the cellular model used.
It is essential to carry out the
staining protocol with Annexin V in Hepes buffer since the binding of this
protein to the phosphatidylserine is calcium dependent, otherwise it is
possible to observe no or a weak increase in Annexin V binding. For this
reason Trypan Blue in the alternative protocol is also diluted in Hepes.
3.4 Anticipated Results
Figure 1 shows the analysis of a
sample of HL-60 control cells (left panel) and of cells treated with 0.15
µM camptothecin for 6 hours (right panel). In the control sample
we observe a population of Trypan Blue positive and FDA negative cells
(Nec, 6%) and a population of apoptotic cells with weak green fluorescence
(2nd-3rd log decade) and negative for Trypan Blue (Apo, 6%). Both of these
populations are normally found in culture. The live cells however, show
a strong green fluorescence (4th log decade) (L; 87%). When the cells are
exposed to an apoptotic trigger, it is possible, whilst maintaining the
same analysis quadrants and the same voltages of the photomultipliers,
to observe the modifications which occur in the various subpopulations,
previously identified (L, Nec, Apo). In fact, we observe the increase of
a few percent in the necrotic population, (Nec, 8%) whilst the FDA dim
population, decreasing its fluorescence gradually, increases six-fold (Apo,
36%) and simultantaneously the numeric value of the live population decreases
(L, 56%). In Figure 2, the expression of LFA-1 antigen is analyzed in the
same population tested for FDA, using a CD11a antibody conjugated with
phycoeritrine. In the panel A figure, the expression of the antigen in
the control population is shown (lighter line). It is possible to note
that almost 95% of control cells is LFA-1 positive (negative cells are
dead or apoptotic, data not shown). In the same hystogram we also show
the expression of this antigen in the camptothecin-treated HL-60 cells
after 6 hours (darker line). In these cells, a strong decrease in the antigen
expression occurs. In panel B the antigen expression in the whole camptothecin
treated HL-60 cells is shown (lighter line) together with the expression
observed only in the apoptotic population (dim FDA, R1 = lower left quadrant,
Fig. 1). It is evident that apoptotic cells lost this antigen during apoptotis.
If we use Annexin V binding to identify the apoptotic population, we obtain
a similar result by drawing a gate around the Annexin V positive - Trypan
Blue negative population (Figure 3).
Figure 1
Figure 2
Figure 3
3.5 Time Considerations
Both protocols, basic
and alternative, are very quick to carry out. In both, the cell preparation
takes about 20 minutes, antibody binding takes another 30 minutes. The
FDA staining needs 5-10 minutes while the Annexin V binding takes 15 minutes.
The Trypan Blue staining takes just few seconds but we suggest a time of
5 minutes to stabilize the number of stained cells (dead cells). The total
assay time (between 65 and 70 minutes) should be taken into account when
time course experiments are performed.
3.6 Key References
-
Carbonari M., M. Cibiati, M. Cherchi,
D. Sbarigia, A.M. Pesce, L. dell’Anna, A. Modica and M. Fiorilli. 1994.
Detection and characterization of apoptotic peripheral blood lymphocytes
in human immunodeficency virus infection and cancer chemotherapy by a novel
flow immunocytometric method. Blood 83:1268-1277. Chrest
F.J., M.A. Buchholz,
-
Y.H. Kim, T.K. Known, and A.A. Nordin.
1993. Identification and quantification of apoptotic cells following anti-CD3
activation of murine G0 T cells. Cytometry 14:883-890.
Curnow S.J., M. Barad, M. N. Brun-Roubereau
and A.M.
-
Schmitt-Verhulst. 1994. Flow-cytometric
analysis of apoptotic and non-apoptotic T-cell receptor-transgenic thymocytes
following in vitro presentation of antigen. Cytometry 16:41-48.
-
Darzynkiewicz Z., G. Juan, X. Li, W.
Giorczyca, T. Murakami, and F. Traganos. 1997. Cytometry in cell necrobiology:
analysis of apoptosis and accidental cell death (necrosis). Cytometry
27:1-20.
-
Dransfield I., A.M. Buckle, C. Haslett,
and N. Hogg. 1994. Neutrophil apoptosis is associated with a reduction
in CD16 (FcgRIII)
expression. J.Immunol.153:1254-1263.
-
Dransfield I., S.C. Stock and C. Haslett.
1995. Regulation of cell adhesion molecules and function associated with
neutrophil apoptosis. Blood 85:3264-3273.
-
Frey T. 1997. Correlated flow cytometric
analysis of terminal events in apoptosis reveals the absence of some changes
in some model systems. Cytometry 28:253-263.
-
Homburg C.H., M. de Hass, A.E. von dem
Borne, A.J. Verhoeven, C.P. Reutelingsperger and D. Ross. 1995. Human neutrophils
lose their FcgRIII
and acquire Annexin V binding sites during apoptosis in vitro. Blood
85(2):532-540.
-
Horvatinovich J.M., S.D. Sparks, and
M.J. Borowitz. 1994. Detection of terminal deoxynucleotidyl transferase
by flow cytometry: a three color method. Cytometry (CCC) 18:228-230.
-
Ormerod M.G. 1994. Further application
in cell biology. In: Flow Cytometry: A Practical Approach, Second
Edition, M.G. Ormerod (Ed), IRL at Oxford University Press, pp. 261-273.
-
Renò F., E. Falcieri, F. Luchetti,
S. Burattini and S. Papa. 1997. Discrimination of apoptotic cells in flow
cytometry using Trypan Blue and FDA. Eur. J. Histochem. Vol. 41
Suppl.2:115-116.
-
Schmid I., C.H. Uittenbogaart and J.V.
Giorgi. 1994. Sensitive method for measuring apoptosis and cell surface
phenotype in human thymocytes by flow cytometry. Cytometry 15:12-20.
-
Swat W., K Ignatowicz, and P. Kisielow.
1991. Detection of apoptosis in immature CD4+8+ thymocytes by flow cytometry.
J.Immunol. Methods 137:79-87.
-
Weber G.F., J. Daley, S.K. Kraeft, L.B.
Chen, and H. Cantor. 1997. Measurement of apoptosis in heterogeneous population.
Cytometry 27:136-144.
-
Wu X.M., Z. Ao, M. Hegen, C. Morimoto
and S.F. Schlossman. 1996. Requirement of Fas (CD95), CD45, and CD11a/CD18
in monocyte-dependent of human T cells. J.Immunol.157:707-713.
Appendix
1: Stock solutions
RPMI 1640 supplemented with 10%
heat inactivated fetal calf serum (FSC), Penicillin/Streptomycin 1000U/1000µg/ml,
200 mM Glutamine, Sodium piruvate 100 mM and buffered with NaHCO3.
Phosphate buffer (PBS): 138 mM NaCl,
27 mM KCl, 1.76 mM KH2PO4, 8 mM Na2HPO4
Hepes buffer :10 mM Hepes/NaOH, pH
7.4, 150 mM NaCl; 5 mM Kcl, 1 mM MgCl2, 1.8 mM CaCl2.
Store at 4°C, stable for 1 month.
Fluorescein diacetate (FDA): dissolved
in acetone (1 mg/ml) and diluted in PBS at a concentration of 1µg/ml
shortly before use. Store at 4°C and protect from light. Stable for
at least 2 months.
Trypan Blue: dilute at 0.8 mM in
PBS or Hepes. Store at room temperature. Stable for 1 month. Trypan Blue
is considered a cancerogenic product .
Appendix 2: Reagents
Annexin V/Fitc Cod. BM S306FI
Bender-Medsystems
Camptothecin Cod. C9911
Sigma
CD11a Cod. 45195F / CBL451 PE
Eurobiox
FDA Cod. T-F-7378
Sigma
Trypan Blue Cod. T-8154
Sigma
Appendix 3: Equipment
Centrifuge Sorval, mod.MC12V, DuPont.
Epics XL cytometer, equipped with
an air-cooled argon ion laser tuned at 488nm wavelength, cod.
Coulter Corporation, U.S.A..