Diphenyliodonium, an NAD(P)H Oxidase Inhibitor, also Induces Apoptosis |
Contributors: Nianyu Li, Kathy Ragheb, Gretchen Lawler, J. Paul Robinson |
Affiliation: Purdue University Cytometry Laboratories |
URL: http://flowcyt.cyto.purdue.edu |
The iodonium compounds diphenyleneiodonium (DPI) and diphenyliodonium (IDP) are well-known phagocyte NAD(P)H oxidase inhibitors. However, at high concentrations they could also suppress ATP production by inhibiting mitochondria complex I NADH reductase. Since mitochondria have been shown to play a crucial role in producing superoxide and inducing apoptosis, we therefore investigated the effect of iodonium compounds on mitochondria-derived superoxide and on apoptosis. Undifferentiated HL-60 cells, in which the functional NADPH oxidase is not present, were incubated with various concentrations of iodonium compounds before assay. Superoxide production was studied by flow cytometry using hydroethidine as the fluorescent probe. Inhibition of oxygen consumption and DNA fragmentation was also investigated. Our results showed that superoxide production of HL-60 cells was increasd in one hour with both rotenone and iodonium compounds. The degree of respiration inhibition was not proportional to the amount of superoxide produced. Polyacrylamide gel assay also showed that the production of superoxide induced by both rotenone and iodonium compounds was able to trigger DNA fragmentation at 24 hours. These observations suggest that IDP can induce apoptosis through its effect on mitochondria, possibly by inducing superoxide production. Key words: Apoptosis; Iodonium compounds; HL-60 cells; Flow cytometry Mitochondria are usually known as the site where oxidative phosphorylation occurs in eukaryotes. Recently it was recognized that this organelle played an important role in signal transduction pathways including oxidative stress and apoptosis. Mitochondria are involved in apoptosis on at lease three aspects. First, upon stimulation, cytochrome c is released from mitochondria to the cytosol. Once in the cytosol, cytochrome c can lead to the activation of caspase 3. Cytochrome c release was found during chemotherapeutic reagent or staurosporine treatment, UV irradiation, growth factor withdrawal, Fas ligation (Apo-1, CD95/CD95L), and tumor necrosis factor receptor activation (Adachi et al., 1997; Kharbanda et al., 1997; Kluck et al., 1997; Yang et al., 1997; Bossy-Wetzel et al., 1998). However, in some other cases, mitochondria permeability transition pores can open and release another apoptogenic protein, apoptosis inducing factor (AIF), to the cytosol. The second way in whick the mitochondria to regulate apoptosis is by disrupting the respiratory chain. On the one hand, this respiratory chain is so vital to the mammalian cell that breaking the chain will certainly lead to cell death, apoptosis or necrosis. On the other hand, during apoptosis when cytochrome c is released from mitochondria to the cytosol, the respiratory chain will be broken by the loss of the electron carrier - cytochrome c. Previous studies showed that inhibitors of the respiratory chain were able to induce apoptosis. These inhibitors include the complex I inhibitor rotenone (Marchetti et al., 1996), the complex III inhibitor antimycin A (Marchetti et al., 1996), the complex VI inhibitor cyanide (Pastorino et al., 1993, Shimizu et al., 1996), the complex V inhibitor oligomycin (Castedo et al., 1996), and the protonophore inhibitor mClCCP (Zamzami et al., 1996). Finally, mitochondria are one of the major sites for production of reactive oxygen species are produced. Reactive oxygen species (ROS) usually include superoxide anions, hydrogen peroxide, organic peroxides and radicals. Most of them are small and highly reactive molecules. ROS can be produced in vivo by many enzyme systems including NADPH oxidase, NADH oxidase, xanthine oxidase, 5-lipoxygenase and the mitochondria respiratory chain complexes. Among them, mitochondria respiratory chain is the most important ROS producing enzyme system. The mitochondria-derived superoxide is important not only because mitochondria respiratory chain components are present on all eukaryotic cells, but also because the superoxide produced in mitochondrial can readily influence the mitochondria function without having to diffuse a long way from the cytosol. The production of ROS by mitochondria is usually enhanced when the mitochondria respiratory chain is interrupted. Early reports showed that complex I inhibitor retenone and 1-methyl-4-phenylpyridinium could enhance ROS production in mitochondria (Turrens and Boversis, 1980). However, some recent data showed that this kind of upstream respiratory chain inhibitors could prevent ROS accumulation and thus protect cells from apoptosis (Schulze-Osthoff et al., 1992; Quillet-Mary et al., 1997). Also, the result of chemiluminscence showed that both retenone and diphenyleneiodonium were able to inhibit the ROS production in mitochondria (Li and Trush, 1998). Diphenyleneiodonium (DPI) and diphenyiodonium (IDP) are well known as flavoprotein inhibitors. They have been reported to inhibit phagocyte NADPH oxidase (Cross and Jones, 1986), nitric oxide synthase (Stuehr et al., 1991), xanthine oxidase (Sanders et al.), and P-450 NADPH reductase. They could also inhibit mitochondria respiratory chain complex I NADH reductase. These compounds are widely used to study the protein-bound flavin, to inhibit superoxide production, and to inhibit apoptosis in many instances. However, due to their function on mitochondria respiratory chain complex I, the iodonium compounds also have the potential to increase the production of superoxide and influence mitochondria function. Therefore we investigated the effect of iodonium compounds on superoxide production and furthermore the effect of these chemicals on apoptosis. Collection of neutrophils Peripheral blood was collected from healthy adult volunteers by standard venipuncture methods in a sterile tube containing preservative-free heparin. Preparation of human neutrophils from normal venous blood was performed as previously described (Robinson, 1993). Briefly, fresh venous blood was placed in a sterile 50ml centrifuge tube and diluted with NH4Cl lysing solution (dilution 16.7:1). Tubes were centrifuged at 400 ´ g for 10 minutes at 4ºC. The supernatant was decanted and cells were washed twice with HBSS two times. Hankss balanced salt solution (HBSS) was composed of 137mM NaCl, 5.4mM KCl, 0.7mM, 1.2 mM NaHCO3, 12.2 mM glucose (0.22%), 27.5 mM Tris, 1.87 mM Ca2+ and 0.8 mM Mg2+, adjusted to pH 7.4. The cell concentration was determined by a Coulter counter (Beckman-Coulter Corp., Hialeah, FL).and was adjusted to 2.0 _ 106 cells/mL in HBSS Growth and Differentiation of HL-60 cells HL60 cells were cultured in RPMI-1640 medium supplemented with 10% fetal calf serum (FCS) and 10% new-born bovine serum (NBS). Cells at a density of 1 ´ 106 cells/ml were pushed into differentiation by the addition of 1.3% DMSO. After 3-4 days of culture (72-96 hours), cells were harvested by centrifugation at 250g for 10 minutes. Measurement of Superoxide production Measurement of O2- in neutrophils and HL-60 cells was performed as described before with some modification (Robinson, 1993). Cells were treated with various reagents and collected at different time points. After washing twice with PBS, the cells were resuspended in PBS containing either 10 m M or 1 m M of HE. In the case of ethidium bromide, 10 m M of dye was present. After 10 minutes incubation at 37ºC, the cells were washed and resuspended in PBS. The cell concentration was adjusted to 2 ´ 106 per ml. The cell suspension was placed into 12 _ 75 mm tubes for assay. (Coulter cytometry, Hialeah, FL). All studies were carried out using an air-cooled 15 mW argon laser (Cyonics Model 2201, San Jose, CA) operating at a wavelength of 488 nm. Fluorescent signals were collected by using a 610 nm long pass absorbance filter. Respiration measurement Oxygen consumption was measured with a Clark oxygen electrode (model 5300: Yellow spring Instrument Co., Yellow spring, OH) as described before (Rustin et al., 1994). Briefly, 5 ´ 107 cells were collected and resuspended in a medium containing 0.3 M mannitol, 10 mM KCl, 5 mM MgCl2, 1 mg/ml BSA, 10 mM KH2PO4 and 0.01% digitonin. 1.6 ml of cell suspension was injected into the respiration chamber which was then sealed. Respiration was calculated as the rate of changes in the oxygen concentration after the addition of various mitochondria complex I inhibitors, assuming the initial oxygen concentration as 6.8 mg O2 / liter. DNA fragmentation The DNA fragmentation assay was performed according to the method described previously with some modification (Herrmann et al., 1994) Untreated or treated HL-60 cells were washed twice with PBS (4°C, pH 7.4) and collected by centrifugation at 500g for 5 minutes. Cell concentration was adjusted to 2 ´ 107 / ml. The cell pellets were treated with 0.5 ml lysis buffer (10 mM Tris-HCL, ph7.4, 10 mM EDTA, 0.5% sodium dodecyl sulfate) for 10 minutes on ice. After treated with Rnase A (final concentration 100 m g/ml) for 1hour at 37° C, the cells were incubated at 50° C for 4 hours in the present of 100 m g/ml proteinase K. DNA was then precipitated with 50 m l of 3M sodium acetate (pH 5.2) and 1ml of cold (4° C) 100% ethanol. Finally, DNA was dissolved in Tris-EDTA. For analysis, 10 to 20 m l of DNA was loaded on a 1.2% agarose gel containing 10 m g/ml ethidium bromide. Electrophoresis was performed in 0.5 ´ Tris-boric-EDTA buffer at 70 V for 2 hours. DNA was visualized under ultraviolet light and photographed. The inhibition of superoxide production of NADPH oxidase by iodonium compounds Hydroethidine has been shown to be very useful in the study of cellular oxidative metabolism. A major advantage of hydroethidine is its ability to distinguish between superoxide and hydrogen peroxide (Robinson, 1998). The dye enters the cell freely and reacts specifically with superoxide to produce ethidium bromide. Consistent with previous reports, our result showed that IDP could inhibit superoxide production by NADPH oxidase both in neutrophils and in dimethylsulfoxide (DMSO) differentiated HL-60 cells (Figure 1, 2). Similarly, DPI could also inhibit NADPH oxidase both in neutrophils and in differentiated HL-60 cells (data not shown). Stimulation of superoxide production by iodonium compounds and mitochondria complex I inhibitor rotenone Iodonium compounds are not only NADPH oxidase inhibitors, but also mitochondria complex I NADH reductase inhibitors. It is suspected that they acted on a site between NADH and the Fe-S clusters (Majander et. al, 1994), in contrast to rotenone, whose inhibition site is between the Fe-S clusters and ubiquinone (Palmer et. al, 1968). Since rotenone was reported to be a good candidate to mimic the TNF function on mitochondria complex I, including superoxide production, mitochondria membrane permeability transition, cytochrome c release and apoptosis (Higuchi et. al, 1998), we thus investigated the possibility that iodonium compounds have similar function. Again hydroethidine is used to study superoxide production in the presence mitochondria complex I inhibitors. Previous results on mitochondria derived superoxide production were contradictory. Early reports showed that rotenone and 1-methyl-4-phenylpyridinium could enhance ROS production both in intact mitochondria and in submitochondrial particles (Turrens and Boversis, 1980). Dichlorodihydrofluorescein derivatives or dihydrorhodamine-123 based assay also showed that rotenone could enhance mitochondrial ROS production. However, recent chemiluminscence data showed that both rotenone and diphenyleneiodonium were able to inhibit ROS production in mitochondria (Li and Trush, 1998). Our result clearly showed similarly to rotenone, both DPI and IDP could induce superoxide on undifferentiated HL-60 cells. It was found that 10 nM to 1 mM rotenone could induce superoxide production through a dose-dependent manner (Figure 3). Higher concentration of rotenone could not further increase superoxide production. Similarly, DPI or IDP could induce superoxide production in a concentration range of 1 to 100 m M and 10 to 200 mM, respectively (Figure 4, 5). Superoxide levels did not increase after one hour (Figure 6, only IDP data are shown). Hydroethidine is widely used to monitor superoxide production in intact cells and isolated mitochondria. However, it has also been reported that mitochondrial membrane potential could influence the distribution of ethidium bromide across mitochondria (Rottenberg et. al, 1984). The enhancement of ethidium fluorescence could indicate loss of mitochondria membrane potential instead of production of superoxide (Budd et al., 1997). It was suggested that a low concentration (1 mM) of hydroethidine should be used in place of a relative high concentration (10 mM). To evaluate our results on superoxide production by rotenone and iodonium compounds, a parallel experiment was performed in which HL-60 cells were loaded with 10 to 25 mM ethidium bromide. Flow cytometry showed when 25 mM ethidium bromide was used, basal ethidium fluorescence of HL-60 cells are very close to the basal fluorescence in HL-60 cells which were incubated with 10 mM hydroethidine (Figure 7), indicating that intracellular ethidium bromide concentrations in both populations were very close. However, it was found that 50 to 200 mM IDP could not induce a fluorescence enhancement in both 10 and 25 mM ethidium bromide loaded cells, in contrast to 10 m M hydroethidine loaded cells (Figure 8A, 8B). This result indicated that the increasing of ethidium bromide fluorescence induced by IDP was due to superoxide production. IDP treated HL-60 cells were also studied by incubating with 1 mM of hydroethidine. The result showed that as with 10 mM hydroethidine loaded cells, IDP was still able to enhance ethidium fluorescence in a dose-dependent manner (figure 9). Combined all these observation, our conclusion is that iodonium compounds could induce superoxide production on HL-60 cells, possibly through its inhibition effect on mitochondria respiratory chain complex I. Oxygen consumption The mitochondria complex I impairment by rotenone and iodonium compounds was estimated by measuring cell respiration. Although all of these three compounds were able to inhibit the mitochondrial respiratory chain at complex I, they seemed to have different potency. Cell respiration was inhibited to 65% of control by 10 nM rotenone and to less than 50% (44%) of control by 20 nM Rotenone (Figure 10-A). Higher concentration of rotenone (500 nM to 1 mM) could entirely block cell respiration (inhibition >95%). In contrast to rotenone, iodonium compounds inhibit cell respiration less effectively. DPI could inhibit cell respiration to 60% at 100 mM and 43% at 200 m M (Figure 10-B). Large concentration of DPI (1 mM) could inhibit cell respiration to less than 20%. Like to DPI, IDP only inhibited cell respiration at high concentrations, to 34% by 1 mM and 28% by 2 mM (Figure 10-C). Rotenone could produce more superoxide at 1mM than at 100nM while the degree of cell respiration inhibition at these concentrations was close. However, 100 m M DPI and 1 mM DPI produced same amount of superoxide while their cell respiration inhibitory effect was different. Similarly to DPI, 200 mM DPI and 2 mM DPI produced same amount of superoxide while inhibiting respiration at different degrees. It was concluded that the degree of inhibition of cell respiration had no direct relationship with the amount of superoxide produced. DNA fragmentation Rotenone was reported to induce apoptosis on various cell lines. Using a DNA fragmentation technique we also found that 1 mM rotenone could trigger DNA fragmentation at 24 hours while 100 nM rotenone could not (Figure11). It was also found that both 100 m M DPI and 200 mM IDP could trigger DNA fragmentation at 24 hours (Figure 11). It was found that rotenone and iodonium compounds could induce apoptosis only at concentrations that sufficient superoxide was produced. The degree of cell respiratory inhibition did not reflect the ability to induce DNA fragmentation. It was also found that the DNA fragmentation induced by these mitochondria complex I inhibitors could be inhibited by the non-specific caspase inhibitor Z-VAD, indicating that caspase is involved in these process. Conclusion Diphenyleneiodonium and diphenyiodonium, despite their inhibitory effect on many superoxide-producing enzymes, could also enhance superoxide production, possibly by inhibiting mitochondria complex I. And this enhanced superoxide production could induce apoptosis on HL-60 cells. Adachi S, Cross AR, Babior BM, Gottlieb RA. Bcl-2 and the outer mitochondrial membrane in the inactivation of cytochrome c during Fas-mediated apoptosis. J. Biol. Chem. 1997; 272:21878-21882 Bossy-Wetzel E, Newmeyer DD, Green DR. Mitochondrial cytochrome c release in apoptosis occurs upstream of DEVD-specific caspase activation and independently of mitochondrial transmembrane depolarization. EMBO J. 1998; 17:37-49 Budd SL, Castilho RF, Nicholls DG. Mitochondrial membrane potential and hydroethidine-monitored superoxide generation in cultured cerebellar granule cells. FEBS letters. 1997; 415; 21-24 Castedo M, Hirsch T, Susin SA, Zamzami N, Marchetti P, Macho A, Kroemer G. Sequential acquisition of mitochondrial and plasma membrane alterations during early lymphocyte apoptosis. J. Immunol. 1996; 157:512-521 Cross AR, Jones OT. The effect of the inhibitor diphenylene iodonium on the superoxide generating system of neutrophils. Specific labelling of a component polypeptide of the oxidase. Biochem J. 1986; 237(1):111-6 Higuchi M, Proske RJ and Yeh ET. Inhibition of mitochondrial respiratory chain complex I by TNF results in cytochrome c release, membrane permeability transition, and apoptosis. Oncogene 1998; 17: 2515-2524 Kharbanda S, Pandey P, Schofield L, Israels S, Roncinske R, Yoshida K, Bharti A, Yuan ZM, Saxena S, Weichselbaum R, Nalin C, Kufe D. Role for Bcl-xL as an inhibitor of cytosolic cytochrome C accumulation in DNA damage-induced apoptosis. Proc. Natl. Acad. Sci. U.S.A. 1997; 94:6939-6942 Kluck RM, Bossy-Wetzel E, Green DR, Newmeyer DD. The release of cytochrome c from mitochondria: a primary site for Bcl-2 regulation of apoptosis. Science 1997; 275:1132-1136 Li Y, Trush MA. Diphenyleneiodonium, an NAD(P)H oxidase inhibitor, also poetnely inhibits mitochondrial reactive oxygen species production. Biochem. Biophy. Res. Comm. 1998; 253:295-299 Majander A, Finel M, Wikstrom M. Diphenyleneiodonium inhibits reduction of iron-sulfur clusters in the mitochondrial NADH-ubiquinone oxidoreductase (Complex I). J. Biol. Chem. 1994; 269: 21037-21042 Marchetti P, Hirsch T, Zamzami N, Castedo M, Decaudin D, Susin SA, Masse B, Kroemer G. Mitochondrial permeability transition triggers lymphocyte apoptosis. J. Immunol. 1996; 157:4830-4836 Palmer G, Horgan DJ, Tisdale H, Singer TP, Beinert H. Studies on the respiratory chain-linked reduced nicotinamide adenine dinucleotide dehydrogenase. XIV. Location of the sites of inhibition of rotenone, barbiturates, and piericidin by means of electron paramagnetic resonance spectroscopy. J Biol. Chem. 1968; 243(4): 844-7 Pastorino JG, Snyder JW, Serroni A, Hoek JB, Farber JL. Cyclosporin and carnitine prevent the anoxic death of cultured hepatocytes by inhibiting the mitochondrial permeability transition. J. Biol. Chem. 1993; 268:13791-13798 Quillet-Mary A, Jaffrezou JP, Mansat V, Bordier C, Naval J, Laurent G. Implication of mitochondrial hydrogen peroxide generation in ceramide-induced apoptosis. J. Biol. Chem. 1997; 272:21388-21395 Robinson, J.P. (1993) "Handbook of Flow Cytometry Methods". Wiley-Liss, NY Robinson J.P. (1998) Oxygen and nitrogen reactive metabolites and phagocytic cells. In "Phagocyte Function: a Guide for Research and Clinical Evaluation." Robinson JP and Babcock GF, eds. NY, Wiley-Liss Rottenberg H. Membrane potential and surface potential in mitochondria: uptake and binding of lipophilic cations. J Membr. Biol. 1984; 81:127-38 Rustin P, Chretien D, Bourgeron T, Gerard B, Rotig A, Saudubray JM, Munnich A. Biochemical and molecular investigations in respiratory chain deficiencies. Clin Chim Acta 1994; 228: 35-51 Sanders SA, Eisenthal R, Harrison R. NADH oxidase activity of human xanthine oxidoreductase--generation of superoxide anion. Eur J Biochem 1997; 245(3):541-8 Schulze-Osthoff K, Bakker AC, Vanhaesebroeck B, Beyaert R, Jacob WA, Fiers W. Cytotoxic activity of tumor necrosis factor is mediated by early damage of mitochondrial functions. Evifactor is mediated by early damage of mitochondrial functions. J. Biol. Chem. 1992; 267:5317-5323 Shimizu S, Eguchi Y, Kamiike W, Waguri S, Uchiyama Y, Matsuda H, Tsujimoto Y. Retardation of chemical hypoxia-induced necrotic cell death by Bcl-2 and ICE inhibitors: possible involvement of common mediators in apoptotic and necrotic signal transductions. Oncogene 1996; 12:2045-2050 Stuehr DJ, Fasehun OA, Kwon NS, Gross SS, Gonzalez JA, Levi R and Nathan CF. Inhibition of macrophage and endothelial cell nitric oxide synthase by diphenyleneiodonium and its analogs. FASEB J 1991; 5: 98-103 Turrens JF and Boveris A. Generation of superoxide anion by the NADH dehydrogenase of bovine heart mitochondria. Biochem. J. 1980; 191:421-427 Yang J, Liu X, Bhalla K, kKim CN, Ibrado AM, Cai J, Peng TI, Jones DP, Wang X. Prevention of apoptosis by Bcl-2: release of cytochrome c from mitochondria blocked. Science 1997; 275:1129-1132 Zamzami N, Susin SA, Marchetti P, Hirsch T, Gomez-Monterrey I, Castedo M, Kroemer G. Mitochondrial control of nuclear apoptosis. J. Exp. Med. 1996; 183:1533-1544 ![]() Figure 2 ![]() Figure 3 ![]() Figure 4 ![]() Figure 5 ![]() Figure 6 ![]() Figure 7 ![]() Figure 8a ![]() Figure 8b ![]() Figure 9 ![]() Figure 10 ![]() Figure 11 ![]() |
![]() |
![]() |