QC TRACKER

 

Automated Software for the daily monitoring

of the performance of a flow cytometer

Introduction

 

Clinical or research use of flow cytometry instrumentation requires that the cytometer be demonstrably proven to provide reproducible results. The QC Tracker software is a program from Phoenix Flow Systems which turns any personal computer into a Quality Control(QC) monitoring system for a flow cytometer. This QC program is designed to aid an operator in detecting and monitoring drift in the performance of a flow cytometer (or multiple cytometers). Stability in the QC data indicates that the data produced by the instrument are not the result of instrument artifact. Conversely, instability indicates the necessity for intervention by either the service personnel or the primary operator.

 

How the QC Tracker software works

 

The QC Tracker software reads the information stored in the Flow Cytometry Standard (FCS)1 1.0/2.0 data files and Coulter Profile™ data files generated by a flow cytometer. The information is stored in a QC database so that it can be retrieved and graphed for comparison with previous information from the same flow cytometer. The following information is stored in the QC Tracker Database:

 

 

In addition, QC Tracker will calculate the mean channel and standard deviation of all of the databased information for each parameter of the first 10-30 (user selectable) cytometer data files that are entered into the QC Tracker database. Once the first 10-30 data files have been read into the QC database, the Means and standard deviations are calculated and locked until a service event or the particle lot number is changed. (If either of these two events is logged into the QC Tracker software, the standard deviations are recalculated and locked again after 10-30 new QC data files.)

 

After the initial 10-30 QC data files, QC Tracker uses ± 2 Standard Deviations from the mean channel as the acceptable range criteria for rendering a pass/fail decision on whether the flow cytometer passes QC each day and displays this information on a Daily Report. See figures 1 and 2.

 

Figure 1-Passing Daily QC Report

Figure 2-Failed QC Report-Note on the Failed QC Report, the parameters which failed are high lighted.

 

The QC Tracker software can also display and calculate Levi-Jennings graphs for any of the above-mentioned databased information. The user can query the database to generate Levi-Jennings graphs for 15, 30, 60, 90, 180 or 360-day time spans. See figure 3.

 

 

Figure 3-Levi-Jennings Report

 

In addition, the program includes a database for tracking and printing operator comments on changes made in the daily QC monitoring of the flow cytometer, i.e., the lot number of particles used for calibration, service calls performed on the flow cytometer or any other operator comments pertaining to the performance of each individual cytometer. See figure 4.

 

 

Figure 4-Comment Report

 

User comments can include items such as:

 

The QC Tracker software can monitor multiple cytometers or types of alignment particles at one time. This means the software can monitor the particles used for alignment, such that the PMT voltages are held constant and the change in the mean channel of the particles is tracked. Or, the software can also monitor the particles used for setting fluorescence intensities such that the mean channel of the particles is held constant, and the PMT voltages are adjusted to keep the particles in their target channels. And, the software can also monitor chick red blood cells or a DNA standard.

 

The only limitation in the software is there must be only one major peak in the data collected.

Setting up the flow cytometer to use the QC TRacker Software

 

It will be much easier to install and use the QC Tracker software if the flow cytometer has been set up first prior to installation of the software in the following manner and a few QC data files have been generated.

 

Type of Particle or Standard to USE

 

Any particles or standard can be monitored by the QC Tracker software provided they contain only one peak, the peak generated is gaussian and the peak falls in the middle 80% of the histogram. Double humped peaks are unacceptable. The cytometer should be repaired or a different particle used before using the QC Tracker software. Phoenix Flow Systems recommends the Rainbow Particles from SpheroTech, catalogue #RFP-30-SA. Phone: (847)680-8922 or available from Phoenix Flow. (A free sample is included with the QC Tracker software.) An adequate quantity should be purchased that will last at least 6 months.

 

ACQUISITION PROTOCOL

 

Once the flow cytometer is turned on and warmed up, the QC particles should be run every day to access the performance of the instrument. A protocol to acquire each type of QC particle data must be set up on the flow cytometer for each type of QC particle used with the following criteria:

 

The two types of Particles used for daily QC

 

Normally, two different types of particles should be run every day to check the performance of the flow cytometer: Alignment Particles and Fluorescence Standardization Particles. Alignment particles check the basic performance of the flow cytometer, i.e. are the PMT's, laser, electronics and fluidics functioning correctly. Fluorescence standardization particles have fluorescence intensities similar to cells and are used to check the performance of the flow cytometer at the normal operating conditions when analyzing a real sample. They are in a sense, a second level of QC for monitoring the performance of the flow cytometer.

 

Running alignment particles

 

When running alignment particles, the QC Tracker software will monitor the drift in the peak of the QC particles. Operate the amplifiers on the flow cytometer in linear mode. Make sure the alignment particles peak falls above the first and below the last 10% of the histogram. Do not adjust the amplifier and PMT gains from day to day. Make sure the amplifier and PMT gains are saved in the cytometer protocol and are the same values each time the QC particles are run. Store this QC data everyday and read it into the QC Tracker software to generate a daily QC report on the performance of the flow cytometer.

 

Running fluorescence standardization particles

 

Adjust the amplifiers (log or linear) and PMT’s to the appropriate gains used for running immunofluorescence samples. Run the appropriate particles on the cytometer. The particles must have only one primary peak. Make sure the peak falls above the first and below the last 10% of the histogram. Record the mean channel of the particles in each parameter and keep this for handy reference. Save the QC particle data to its unique directory. The next day, noting the mean channels recorded from the previous day, run the QC particles. Adjust the gain of the PMTS of the flow cytometer so the particle peak comes up in the target channel means from the previous day. Store this data each day to the same QC data file directory (which is different from the alignment particle directory mentioned before). and read it into QC Tracker software to generate a daily QC report on the performance of the flow cytometer.