The following protocol describes the method for measuring apoptosis in the positive and negative control cells that are provided in the APO-DIRECTTM kit. The same procedure should be employed for measuring apoptosis in the cell specimens provided by the researcher. |
Staining SOLUTION COMPONENT |
1 ASSAY |
6 ASSAYS (2 controls + 4 unknown) |
12 ASSAYS (2 controls + 10 unknown) |
---|---|---|---|
TdT Reaction Buffer(green cap) TdT Enzyme(yellow cap) Fluorescein-dUTP(orange cap) Distilled H2O |
10.00 µl 0.75 µl 8.00 µl 32.00 µl |
60.0 µl 4.5 µl 48.0 µl 192.0 µl |
120.0 µl 9.0 µl 96.0 µl 384.0 µl |
Total Volume |
50.75 µl |
304.5 µl |
609.0 µl |
The appropriate volume of Staining Solution to prepare for a variable number of assays is based upon multiples of the component volumes combined for 1 Assay. Mix only enough DNA Labeling Solution to complete the number of assays prepared per session. The DNA Labeling Solution is active for approximately 24 hours. |
NOTE: The Staining Reaction can also be carried out at 22-24C overnight for the control cells. For samples other than the control cells provided in the kit, incubation times at 37C may need to be adjusted to longer periods of time. |