I don't know much about the FACStarPLUS and the FACS VAntage, but if the
signals are not ratioable using the hardware, there is software available
to perform the ratios. Phoenix Flow's LIST-VIEW will ratio any 2 list mode
parameters. This would certainly avoid any problem of coincident particles
through the different LASERs. I'm also sure that there are other software
packages that will perform the same function.

In message Fri, 20 Aug 1993 16:48:59 PST,
Mike_Lance@bdis.com (Mike Lance) writes:

> Subject: Time:4:38 PM
> OFFICE MEMO Ratio on FACS Vantage/FACStar+ Date:8/20/93
>
> The FACS Vantage and the FACStarPLUS are both highly adaptable. However,
> ratioing the delayed signals between two lasers is not possible on either
> system. The suggestion of using the FL2 signal cable to "trick" the
> instrument will result in signals from two events being ratio'ed, not two
> parameters from a single cell being ratio'ed which is probably what you
> want. This is because of the timing of when the signal is ratio'ed, which
> occurs right after the peak. Thus, the ratio is determined right after the
> the particle goes through the first laser (on an FL1/FL2 ratio), before it
> is even seen by the second laser! So if you "trick" the system this way,
> the ratio is actually the ratio of an FL1/"FL2" event that occurred at the
> same time as an actual "FL4" event.
>
> Angela Schultz
> Product Manager, FACS Vantage
{---------------------------------------------------------------------}