Re[2]: 1.Instrument settings 2.saponin

MCosgrove@bdis.com
Wed, 05 Jul 95 16:54:32

Mikael,
Dr. Martz is correct, the FCS list mode files produced by version 1.0 of
CELLQuest do not contain instrument settings, I am sorry if this has caused you
any problems. Version 1.1.1, which is now shipping to FACSCalibur customers,
does write a complete set of instrument settings to the list mode file. The
instrument settings can now be recalled and downloaded from either a list mode
file or a separate instrument settings file. An update for your 1.0 version of
CELLQuest will be available in the near future.

Michael Cosgrove
CELLQuest Project Manager
BDIS

_______________________________________________________________________________

In message Tue, 04 Jul 1995 19:14:57 +0200,
Mikael Vestberg <Mikael.Vestberg@inflam.lu.se> writes:

> Hello everyone,
> Does anyone know how to get instrument settings from a datafile
> collected with Cellquest on the Macintosh? It was possible with the HP
> system and both should be FCS2 standard, shouldn't they?

Here is an excerpt from the on-line help of my free program MFI:

[BEGIN QUOTE]
CYTOMETER/SOFTWARE FILES TESTED
[...]
FACStation 1.0 'CellQuest' MacIntosh System Software 7.1.0 (FACScan) ***
[...]
*** The FACStation 1.0/MacIntosh files tested do not include instrument
gains, PMV's, threshold, or compensation. Becton-Dickinson promises to
rectify this serious deficiency in a future upgrade. This deficiency
precludes MFI from detecting inadvertant changes in instrument settings.
[END QUOTE]

The above statement is based on my September, 1994 correspondence with
Michael Cosgrove, B-D's project manager for CELLQuest. I have CC'd this
message to Mr. Cosgrove -- perhaps he can let us know when the upgrade can
be expected?

/*- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
Eric Martz, Professor of Immunology emartz@microbio.umass.edu
Dept Microbiology Voice: 413-545-2325 FAX: 413-545-1578
Morrill IVN 203, Box 35720, Univ Massachusetts, Amherst MA 01003-5720
- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -*/