FLUORESCENCE MICROSCOPY ANALYSIS OF NUCLEAR ALTERATIONS
DURING APOPTOSIS
Ildo Nicoletti & Roberta Mannucci
Istituto di Medicina Interna e Scienze Oncologiche
Università di Perugia, 06122 Perugia, Italy
EMail: imiso@unipg.it
Introduction
Cells undergoing apoptosis display typical features, namely cell shrinkage,
membrane blebbing, chromatin condensation and nuclear fragmentation. Dramatic
changes occurr within the nucleus during apoptotic death. It is commonly
thought that the nuclear changes are due to activation of endogenous nuclease(s)
which cleaves DNA into oligonucleosomal fragments. This is associated with
the appearance of dense, crescent-shaped chromatin aggregates which line
nuclear membrane. Later, the nucleulus disintegrates, nuclear membrane
develops deep invaginations and, ultimately, the nucleus fragments into
dense granular particles (apoptotic bodies).
Chromatin condensation, nuclear shrinkage and formation of apoptotic
bodies can easily be observed under fluorescence microscopy, after appropriate
staining of nuclei with DNA-specific fluorochromes. Both adherent cell
lines and cells growing as suspension can be examined. However, due to
slight differences in the fixation procedures, two separate methods will
be described. A short paragraph dealing with fixation requirements for
confocal analysis is also included.
.
Protocol for Fluorescence Microscopy: Fixation
1. Adherent cell lines
Materials
-
PBS pH 7.2 (stock solution 2X)
-
Paraformaldehyde 3.7% (A1)
-
Absolute Methanol (or Ethanol)
(A1) Preparation of paraformaldehyde (PFA) solution (100 ml):
-
dissolve 3.7 g of paraformaldehyde in 50 ml H2O, add 2 drops
of 10N NaOH, heat 30 min at 60°C. Add 50 ml of PBS 2X. Adjust pH to
7.2
-
The PFA solution must be freshly made.
-
Attention PFA is toxic if inhaled.
Methodology
-
Wash cells in PBS pH 7.2
-
Fix cells in 3.7% paraformaldehyde (PFA) in PBS pH 7.2 for 15 min. at room
temperature
-
Wash once with PBS for 5 min at room temperature
-
Cover the cells with methanol 5 min. room temperature
-
Wash cells in PBS pH 7.2
-
Incubate with staining solution
Alternative fixation methods:
-
3.7% PFA and 0.2% TritonX-100 in PBS for 5 min at room temperature
-
3.7% PFA and 0.2% SDS in PBS for 5 min at room temperature
-
Methanol for 5 min at room temperature and Acetone for 2 min. at -20°C
* These alternative methods should
be used when the permeabilization of cells (i.e. for a combined immunofluorescence
staining) is a critical point of the procedure
2. Cells growing in suspension
Materials
-
PBS pH 7.2
-
Paraformaldehyde 3.7% (A1)
-
Absolute Methanol (or Ethanol)
Equipment
-
Cytocentrifuge
-
Centrifuge
Methodology
-
Wash cells in PBS pH 7.2
-
Spin the cells down (2x105 cells 5 min. at 500g)
-
Air dry
-
Restart from point 2 of previous paragraph
Protocol for Fluorescence Microscopy: Staining
Nuclei can be stained with different fluorochromes, depending
on experimental requirements. For example, if a simultaneous analysis of
membrane antigens with FITC and/or TRITC is performed, nuclei should be
stained with a UV-excited, blue-emitting fluorochrome such as DAPI (excitation
358 nm, emission 461 nm). Propidium iodide (excitation 536 nm, emission
617 nm) can be used for routinary analysis of nuclear morphology or in
combination with green-emitting fluorochromes (i.e. FITC).
DAPI Staining
Materials
-
4’,6-diamidine-2’-phenylindole dihydrochloride (DAPI ) (A1)
-
Citric acid
-
Tween 20
-
Na2HPO4
Equipment
-
Epifluorescence microscope with appropriate
filter set
(A1) Preparation of DAPI stock solution:
-
Dissolve 1mg /ml of powder in water and make appropriate aliquots that
can be stored at -20°C
Suspension cells
1) Pretratment solution
Citric acid 4.2 g (0.2M)
Tween 20 0.5 ml
Distilled H2O 100 ml
2) Staining solution
Distilled H2O 100 ml
Na2HPO4 7.2 g
DAPI 0.2 mg (200 m l from stock solution)
Methodology:
-
Resuspend cells in 1 vol of pretreatment solution and incubate 5 min at
room temperature
-
Add 9 vol of staining solution and incubate 5 min
-
Put cells into the slide and observe under fluorescence microscope
Adherent cells: 1) Pretreatment solution
Citric acid 2.1 g
Tween 20 0.5 ml
Distilled H2O 100 ml
2) Staining solution
Citric
acid 11.8 g
DAPI 0.2 mg
(200 m l from stock solution)
Distilled H2O
100 ml
Methodology:
-
Put on cells 1 Vol of pretratment solution and incubate 5 min at room temperature
-
Add 6 Vol of staining solution and incubate 5 min
-
Remove the excess of liquid and observe under fluorescence microscope
* Staining solutions can be stored
at room temperature for several weeks protected from light
Support Protocol ( Short Protocol)
Materials
-
DAPI staining solution ( 10 m l from stock solution
in 100 ml of PBS)
Methodology:
-
Wash once cells with DAPI staining solution for 5 min at room temperature
and observe directly under fluorescence microscope (this short method is
suitable for nuclear counter-staining in immunofluorescence analysis)
Propidium Iodide Staining
Materials
-
Propidium Iodide Stock solution 1 mg/ml in PBS (A1)
-
PBS pH 7.2
Equipment
-
Epifluorescence microscope with appropriate filter set
(A1) Preparation of Propidium Iodide stock solution:
-
Dissolve 1mg /ml of PI powder in PBS.
Methodology:
Suspension Cells
-
Resuspend cells in 1 ml PBS
-
Add 10 ml PI stock solution
-
Incubate 10 min at room temperature
-
Examine under the fluorescence microscope
Adherent Cells
-
Cover cells with the appropriate volume of PI working solution (10 mg/ml
PI in PBS).
-
Wash with PBS and mount.
-
Observe under the microscope
Protocol for Confocal Fluorescence Microscopy: Fixation
Confocal laser scanning microscopy (CLSM) has recently emerged as a
technique which offers several advantages over conventional fluorescence
microscopy. The out-of-focus blur is virtually absent from confocal images,
giving the capability for serial optical sectioning of intact specimens
and subsequent three-dimensional reconstruction. For a precise analysis
of stereo-spatial relationships of different cellular organelles, it is
critical that fixation and mounting of specimens preserve the structural
architecture of the cells. The ideal fixative should penetrate tissues
quickly, act rapidly and preserve the cellular structure before the cell
can react to produce structural artifacts. The fixation methods described
below have been optimized for CLSM analysis of apoptotic cells and nuclei.
1. Cells growing in suspension
Materials
-
PBS pH 7.2
-
Paraformaldehyde 3.7% in CSK (A1)
-
0.2% TritonX-100
-
Poly-L-lysine stock solution (mw 300000, 0.01% in water) (A2)
Equipment
(A1) Preparation of PFA in CSK solution (100 ml):
-
Dissolve PFA as above, add 10 ml CSK 10X.
-
Adjust pH to 7.4
-
10X CSK: 100 mM Pipes pH 6.8 (adjust with 10N NaOH)
-
1 M NaCl
-
30 mM MgCl2
-
20 mM EDTA
-
3 M sucrose
* Prepare 10X CSK solution w/o sucrose
(can be stored at room temperature for several weeks).
* The PFA solution with sucrose should
be made freshly
(A2) Preparation of poly-L-lysine coated coverslips:
-
Wash with soap
-
Rinse with distilled H2O
-
Immerse coverslip in absolute ethanol
-
Immerse coverslip in acetone
-
Air dry
-
Make a ring with DAKO-pen around the coverslip edge
-
Put on the poly-L-lysine solution for 30 min at room temperature
-
Pour off exceeding poly-L-lysine
-
Wash with distilled H2O
-
Air dry
* The polyl-L-lysine coated coverslips
can be stored at room temperature for several days
* Store the poly-L-lysine stock
solution at -20°C
Methodology
-
Centrifuge cells at 1200 rpm for 5min.
-
Wash freshly harvested cells (3-5 X 105) in PBS
-
Resuspend cell pellet in 300 m l of PBS
-
Put on poly-L-lysine coated coverslip 30 min at 4°C (A2)
-
Remove (gently!!) unbound cells
-
Wash once with PBS for 5 min at room temperature
-
Fix in 3.7% PFA in CSK for 10 min at room temperature
-
Wash three times in CSK w/o sucrose for 5 min at room temperature
-
Add 0.2% Triton X-100 in CSK w/o sucrose for 5 min at room temperature
-
Wash once in PBS
-
Incubate with staining solution
2. Adherent cell lines
Materials
-
PBS pH 7.2
-
Paraformaldehyde 3.7% in CSK (A1)
-
0.2% TritonX-100
Equipment
Methodology:
-
Growth cells on coverglass and start from point 7 of previous paragraph
Protocol for Confocal Fluorescence Microscopy: Staining
As reported in the previous chapter, in this case too nuclei can be
stained with different fluorochromes, depending on experimental demands.
Since the majority of CLSMs are equipped with argon (488 nm excitation
line) or argon-kripton (488, 568 and 647 nm excitation lines) only fluorochromes
with appropriate excitation spectra can be used. Propidium iodide or long-wavelenght
nuclear fluorochromes (TO-PRO-3, TOTO-3 from Molecular Probes) are both
suitable for revealing nuclear alterations during apoptosis. The staining
method is the same reported in the previous paragraph. After staining samples
should be properly mounted for CLSM processing:
-
Air dry the sample
-
Apply four posts of nail polish (to avoid that coverslip touch the sample)
-
Apply the appropriate volume of mounting medium (A1) (about 10-20 ml,
depending on the height of the specimen)
-
Seal the coverslip with nailpolish
-
Observe under CLSM
(A1) Mounting Medium: there are a number of commercially available
preparations (FluoroGuard from BioRad, SlowFade and ProLong from Molecular
Probes) which contain anti-fade agents to reduce fluorescence photo-bleaching.
A good semi-permanent mountant can be obtained as follows:
-
Make a solution of 20 g polyvinyl alcohol of average mw 10000 (Gelvatol,
Monsanto Chemical Co or Airvol 205, Air Products Ltd) in 80 ml of 140 mM
NaCl, 10 mM Sodium Phosphate buffer, pH 7.2 by stirring for 16 h or heating
in a boiling waterbath
-
Add 40 ml glycerol and str for a further 16 h, then centrifuge at 12000
to remove any undissolved solid. Remove the supernatant, check its pH is
between 6 and 7 and store in 10 ml aliquots in the freezer (-20°C)
-
Add 1 mg DABCO (Sigma) to a 10 ml working solution of this preparation
before use and store at 4°
Protocol for Acridine Orange Staining after DNA denaturation
Acridine Orange (AO) is a metachromatic dye which differentially stains
double-stranded (ds) and single-stranded (ss) nucleic acids. When AO intercalates
into dsDNA it emits green fluorescence upon excitation at 480-490 nm. On
the contrary, it emits red when interacts with ssDNA or RNA. Chromatin
condensation is an early event of apoptosis and the condensed chromatin
is much more sensitive to DNA denaturation than normal chromatin. Therefore,
if RNA is removed by pre-incubation with RNase A and DNA is denaturated
in situ by exposure to HCl shortly before AO staining, apoptotic
cells (which have a larger fraction of DNA in the denaturated form) display
an intense red fluorescence and a reduced green emission when compared
to non-apoptotic interphase cells.
Materials
-
Acridine Orange (A1)
-
Citric acid
-
Na2HPO4
-
Paraformaldehyde
-
PBS
-
DNAse-free RNAse A (A2)
-
Ethanol
-
HCl
Equipment
-
Epifluorescence microscope or Flow cytometer
(A1) Staining solution
-
Acridine Orange 6 mg/ml
-
Citric acid 0.1 M
-
Na2HPO4 0.2 M pH 2.6
-
Prepare 90 ml of citric acid solution, add acridine orange and 10 ml Na2HPO4
* Acridine Orange solution
is stable for several weeks when stored at 4°C and in the dark
(A2) RNAse solution
-
Dissolve 1 mg of RNAse A (use DNAse-free RNase) in 1 ml of distilled water
Methodology:
-
Wash cells (1x106) in PBS and centrifuge at 200 g for 5 min
-
Resuspend the cell pellet in 1 ml PBS
-
Fix cells by transferring the cell suspension in 9 ml 1% paraformaldehyde
in PBS, on ice. Incubate for 15 min on ice
-
Centrifuge at 200 g for 5 min and resuspend the cell pellet in 5 ml PBS,
centrifuge
-
Suspend the cell pellet in 1 ml PBS and transfer the suspension in 9 ml
70% (vol/vol) ethanol, on ice.
-
Incubate for 4 h (The cells can be stored in ethanol for weeks)
-
Centrifuge at 200 g for 5 min and resuspend the cell pellet in 1 ml PBS
-
Add 0.2 ml of RNAse A solution. Incubate at 37°C for 30 min
-
Centrifuge at 200 g for 5 min and resuspend the cell pellet in 0.2 ml PBS
-
Add 0.5 ml of 0.1 M HCl at room temperature
-
After 30-45 sec add 2 ml AO staining solution
-
Observe the cells under fluorescence microscope with an appropriate filter
set.
-
Alternatively, analyse cells by flow cytometry (excitation 488 nm; dot
plot of green fluorescence at 530± 20
nm versus red fluorescence >600 nm).
Key References
-
Hotz MA, F Traganos and Z Darzynkiewicz. 1992.Changes in nuclear chromatin
related to apoptosis or necrosis induced by the DNA topoisomerase II inhibitor
fostriecin in MOLT-4 and HL-60 cells are revealed by altered DNA sensitivity
to denaturation. Exp Cell Res 201:184-191
-
Gorman A, J McCarthy, D Finucane, W Reville and T Cotter. 1994. Morphological
assessement of apoptosis. In: Techniques in Apoptosis. A user’s guide.
TG Cotter and SJ Martin eds, Portland Press, New York, pp. 1-20
-
Pawley JB.1995. Handbook of Biological Confocal Microscopy (2nd
Edition). Plenum Publishing Corporation, New York.