>
> To All:
> In the time course of performing a sort (1-2hrs.) my sorter is experiencing
> drifts in droplet breakoff(as observed by the camera screen on the
> sorter).When this is observed,the sort is stopped and the drop delay/droplet
> breakoff is checked by performing the ususal bead count matrix over several
> drop delays.Typically, the drop delay has in fact changed from its original.
> While I want to hear any and all responses as to what the problem may be(my
> performance or the sorter)assume for the time that there is absolutely no
> clogs,no air/fluidic leaks, and all electronics completely in fine tune.
> Thanks.
> Vincent Falco
> New England Medical Center
> vincent.falco@es.nemc.org
>
If this would happen I would not take the time to do a matrix count. I
would re-calculate the drop delay set the instrument to new values and go.
My sorter is fairly reliable as far as drop delay settings. This sorter
I use now is off by 1 drop. I calculate the DRop delay and add 1. If I
do a matrix sort it tells me the same thing. I do not waste time in the
middle of a sort for a matrix check.
I would be curious if this needs to be done every several hours or just
this once. If it one time a day just reset and go. If it is every 2-3 hrs
then there is another problem.
Go with the flow.
Jim
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web