Re: HIV+ cells/stream-in-air sorting/safety issues

SPERFETTO (sperfetto@hiv.hjf.org)
Thu, 14 Nov 1996 09:22:38 -0500

Frederic,

Awhile ago, after the Lake Placid ISAC meeting, a committee was formed to draft
guidelines for sorting live infectious cells. This effort is headed up by
Ingrid Schmid at UCLA who has done a great job in organizing the information for
other to review. I would suggest an e:mail to her to determine the status of
these guidelines.

schmid@cyto.medsch.ucla.edu
_______________________________________________________________________________
Subject: HIV+ cells/stream-in-air sorting/safety issues
From: Frederic Preffer <preffer@helix.mgh.harvard.edu> at Internet_Gateway
Date: 11/12/96 8:44 PM

Can anyone refer me to any existing regulations/suggestions concerning
safety issues when it comes to cell sorting with UNFIXED samples known to be
from HIV+ patients, or cells purposely infected in vitro? In the absence
of known regulations, what precautions have you implemented in your facility
to insure the safety of the operator? For example, is a BL3 room a minimal,
adequate or more than necessary condition?

thanks
`````````````````````````````````````````````````
Frederic I. Preffer
preffer@helix.mgh.harvard.edu
Department of Pathology- Cox5
100 Blossom St
Massachusetts General Hospital
Boston MA 02114
(617) 726-7481 {fax x2365}
~~~~~~~~~~~~~~~~~~~~~~~~~~~


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu