Re: PI and GFP

David Galbraith (dgalbrai@ag.Arizona.EDU)
Mon, 24 Feb 1997 19:32:17 -0700 (MST)

Wendy:

I think ethanol fixation eliminates GFP fluorescence -- it disrupts the
beta-barrel and lets solvent into the fluorochrome, thereby quenching it.
I'd try other fixatives, such as paraformaldehyde, and try to keep GFP in
a nearly-native environment. Alternatively, maybe
you could get the PI in through electroporation (?)

David Galbraith

On Mon, 24 Feb 1997, Wendy D. Schober wrote:

>
> We are attempting to optimize the procedure to label transfection along
> with cell cycle in HeLa cells. So far we have co-transfected with CD20
> and then labeled the surface with CD20-FITC. We found 5-13% CD20
> positive with good PI patterns after the usual 70% ethanol fixation.
> We used the same cells and fixation procedure with GFP as the reporter and
> hoped to find similar results. Unfortunately, there were few cells GFP
> positive and what may be there are of very low intensity, barely over
> background. We are looking for suggestions for improving our GFP and keeping
> good PI patterns. The GFP is from Invitrogen and they have been somewhat
> helpful, but have not done this with PI.
> Is fixation causing a problem?? If so, is there a better procedure which
> will permeabilize and fix to get both GFP and PI to work?? Ultimately, we
> want this to work in some fibroblast lines, but HeLa is our control.
>
> Thanks in advance for any hints.
> Wendy Schober
> wschober@bcm.tmc.edu
>


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu