Reply to Dr. Farkas...the easiest answer is that these tissues have been
fixed with formalin prior to raraffin embedding. This develops crosslinks which limits accessibility to the DNA by propidium iodide -- therefore, less dye
means less fluorescence, so your peaks will naturally be lower. Several
attempts at eliminating this have been reported -- the best of which I was from
a recent Cytometry article where the authors (I apologize for not remembering
the names...if it's necessary, I'll look it up -- I think it was Overton?) used
heating to elinimate the effect.


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu