Antony Bakke (bakkea@ohsu.EDU)
Thu, 07 Dec 95 10:57:28 PST

Instead of PI in FL2 and Cy3 in FL3, you can use 7AAD in FL3 and PE
in FL2. (see Schmid et al. Cytometry 12:279-285, 1991) However,
7AAD does not give low CVs - usually about 4-6%.

You can also use FITC, PE and PI according to a paper in Cytometry,
but this may require high levels of expression of the antigens
detected with FITC and PE: Corver, et al Cytometry 15: 117-128,
1994.

Tony Bakke (bakkea@ohsu.edu)


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu