-2) PE anti-CD34 reagents always seem to give higher background than FITC (Check
monocytes IgG PE vs. HPCA-2 PE).
-3) We use 100 ul blood + 20 ul Ab.
-4) We have also incubated UCB with mouse serum to lower nonspecific binding.
Ten minutes, RT., then directly mix with antibodies.
-5) Include CD45 FITC and scatter gates to help eliminate small debris. This may
help with UCB, as not all the erythrocytes are lysed.
See The ISHAGE guidelines for CD34+ cell determination by flow cytometry,
Sutherland, Anderson, Keeney, Nayar and Chin-Yee, Journal of Hematotherapy, June
1996.
Dennis
*************************************************************************
* Dennis J. Young Voice : (619) 822-0407 *
* Flow Cytometry Core Facility FAX : (619) 822-0412 *
* University of California, San Diego USA e-mail: djyoung@ucsd.edu *
* http://www-core.ucsd.edu/flow-core.html *
*************************************************************************
______________________________ Reply Separator _________________________________
Subject: staining of cord blood CD34+ cells
Author: yagrawal@messi.uku.fi at @UCSD
Date: 11/21/96 3:52 PM
Dear flowusers,
Problem:I have been having problems with CD34 cell
staining ofwhole Cord Blood(CB). Specifically the Isotype control (IgG1-PE from
B.D) gives non-specific staining (0.2-0.6 % of mononuclear cells). This
small amount of non-specific staining causes difficulty in interpretation
when the CB stained with the specific HPCA-2-PE (also from BD) gives a
positivity in 0.5 % of cells and subtracting the background gives 0%
positivity.
Do others have the same problem, any suggestions/advice is welcome ?.
Methods used:We collect our CB in blood bags (CPDA) and the sample is
stored at 4 C until analyzed which is 1-4 hrs. Staining is according to
standard methods. 50 ul of CB + 20 ul antibody is incubated for 15 min in
cold and in dark. After this 2 ml FACSlysis solution (BD) is added,
incubated for 10 min and centrifuged for 5 min at 400 g. The cells are
then washed twice in PBS and analyzed immediately.
Thanking you all,
Yash Pal Agrawal
Kuopio, Finland
yagrawal@messi.uku.fi
>-- Saved internet headers (useful for debugging)
>Received: from flowcyt.cyto.purdue.edu by mail.ucsd.edu; id MAA05314 sendmail 8
>Received: by flowcyt.cyto.purdue.edu (940816.SGI.8.6.9/930416.SGI.AUTO) for cyt
>Received: from messi.uku.fi by flowcyt.cyto.purdue.edu via SMTP (940816.SGI.8.6
>Received: from localhost by messi.uku.fi (AIX 3.2/UCB 5.64/4.03) id AA
>Date: Thu, 21 Nov 1996 15:52:12 +0200 (WET)
>From: Yashpal Agrawal <yagrawal@messi.uku.fi>
>To: Cytometry M
>Subject: staining of cord blood CD34+ cells
>Message-Id: <Pine.A32.3.95.961121153116.95616B-100000@messi.uku.fi>
>Mime-Version: 1.0
>Content-Type: T
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web