Subject: HIV+ cells/stream-in-air sorting/safety issues
From: Frederic Preffer <preffer@helix.mgh.harvard.edu> at Internet_Gateway
Date: 11/12/96 8:44 PM
Can anyone refer me to any existing regulations/suggestions concerning
safety issues when it comes to cell sorting with UNFIXED samples known to be
from HIV+ patients, or cells purposely infected in vitro? In the absence
of known regulations, what precautions have you implemented in your facility
to insure the safety of the operator? For example, is a BL3 room a minimal,
adequate or more than necessary condition?
thanks
`````````````````````````````````````````````````
Frederic I. Preffer
preffer@helix.mgh.harvard.edu
Department of Pathology- Cox5
100 Blossom St
Massachusetts General Hospital
Boston MA 02114
(617) 726-7481 {fax x2365}
~~~~~~~~~~~~~~~~~~~~~~~~~~~
![]() |
![]() |
![]() |
![]() |
![]() |