Referring to the question of Andreas Schroeder a few points should be
made. The paper in the Journal of Immunological Methods he cites,
states that DiOC(18)/PI canNOT be used for flow cytometric analysis of
cell mediated cytotoxicity; another paper in the same journal (volume
166, p. 45, 1993) demonstrates successful use of this approach.
Association of target and effector cells is a prequisite for cell
mediated cytoxicity to occur. So, one should not be surprised that
it actually does take place. In the flow cytometer one scores doubly
positive (i.e. green/red positive) vs singly positive (only green
positive cells). That takes care of the association of unlabeled
effector cells with labeled target cells.
If one has a lot of clumping in a sample (in particular higher order
clumps), filtering may help. Then, one has to check that there
filtering does not selectively exclude a cell population;
that may lead to a bias in your assay.
Another way of dealing with clumps of particles is by taking into
account that a clump of two spheres which passes by a norrow beam of
light will have a different ratio of pulse width vs pulse height. In
a plot of pulse width vs pulse height, clumps are off the diagonal
(which represents "single" events). Wolfgang Goehde (University of
Muenster, Germany) recognized this phenomenon; it is incorporated in
some flow cytometers, where it is known as the "pulse processor".
Hope this helps,
Martin Poot
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web