We have had good success using the tandem conjugates from several companies
but we have also received a few bum reagents from time to time. Here are a
few suggestions on what to try to check ou the machine.

1. Try checking the alignment with all the filters in using beads. Check to
see that the far red and green maximize response at nearly the same setup.
If not go directly to Moe and do not collect $200.00

2. Try moving your red filter up to a PMT closer to the front using a 99%
reflector in the dichroic slot. This should help identify a problem with
either the PMT or a dichroic in the path.

3. Try out your samples on Larry's machine (Slowly he turns, step by step,
inch by inch ...).

4. Have Moe and Larry come out and buy you a beer.

Good Luck!!

Gary Durack
University of Illinois Biotechnology Center
Voice (217) 244-0559
FAX (217) 244-0466


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu