We have been doing calcium flux on unseparated, stained PBMC. The
protocol we use is pretty standard, with the following important points:

(1) Cells are loaded at 37C with Indo for 45-60 min
(2) Cells are centrifuged and then spun at room temperature in the
presence of 0.02 to 0.1% sodium azide with antibodies. We use only
direct conjugates to avoid crosslinking. It is EXTREMELY important not
to let the cells get too cold (never on ice); they will not flux.
(3) Cells are then washed at room temp 2-3 times to wash out the azide,
and resuspended in medium at room temp. All washes and suspensions are
done in complete growth medium, or just RPMI with 4% serum.
(4) Cells are warmed to 37C for 5 min prior to baseline and stimulation
collections.

I have stained with: CD4, CD8, CD45RA, CD45RO, and L-Selectin. Various
controls have demonstrated that this staining protocol does not
significantly alter the peak calcium, the final calcium, nor the
kinetics--with the possible exception of CD8 causing the CD8 T cells to
flux somewhat better. We will be publishing this shortly in Cytometry.

Finally, I used Consort-VAX to analyze the huge files that arise from
these experiments. (Typically, we collect 8 parameter data at 2000
cells per second for up to 15 minutes). Consort-VAX works quite
efficiently and has shown no problems to date. By the way, I have
rewritten the kinetics analyzer for Consort-VAX to work with files that
have TIME as a parameter rather than storing kinetic information the way
that Consort-VAX does--if anyone is interested, please contact me.

Mario Roederer
Roederer@Darwin.Stanford.Edu


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu