Sheath buffers/H2O/PBS

Joseph Trotter - Salk Institute Flow Cytometry (trotter@flosun.sdsc.edu)
Tue, 16 Nov 93 10:05:27 PST

Flow people,

Be careful about the different refractive indices of sheath buffer
and sample buffer, you can introduce some nice "lens effects" at the
"wrong place" if the indices differ enough. It never hurts to match the
buffers fairly well, especially if scatter cv's are an issue. Running PBS
sample buffers in instruments using water as a sheath buffer sounds like a
formula for obtaining "less than optimal" cv's. We always switch out sheath
buffers to match the sample buffer refractive index if the sample buffer
isn't fairly "close" to our normal PBS buffer. Usually, this is only for
something like chromosomes or plant protoplasts, etc...

Since scatter measurements are largely a measurement of the difference
in refractive index between cells (or whatever) and the medium they are
suspended in, in makes no sense to surround the sample/buffer with a sheath
buffer having yet a different refractive index.

Joe Trotter
Salk Institute


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu