A possible factor in your datafile problem:
I found a group of files which LYSYS II thought were empty (no data)
although the file size and all other indications suggested there were
10,000 cells, 4 parameter by 256 res;
The cause was the operator had included the delimiter character "\"
in the SAMPLE ID text, thus putting the header translation out of synch.
and LYSYS II thought the TOT (number of events) was zero!

I successfully fixed some, but not all, of the files by FACSNet-ing to
a PC and using a DOS text editor to change the offending character in
the file header. Can't remember what editor it wassomething obscure
that can handle huge files.

Some questions that might help isolate the problem:
Does PC-LYSYS handle the files o/k?
What does Cicero use as a text delimiter character?

Best of Luck!

- From Joseph Webster
Centenary Institute, Sydney
cenint04@angis.su.oz.au


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu