Correction: Ab vs. TRITC

ERIC MARTZ (emartz@titan.ucc.umass.edu)
Thu, 17 Dec 92 16:55:08 EST

This is an update and corrected interpretation of my message entitled
'Ab killed by TRITC?' sent 12/16/92.

Direct FACStar titration of the TRITC conjugate gave a half-maximal
binding at a concentration >= 1/25, less than 10% of the input binding
activity. However, today we found that _indirect_ FACScan titration
gives 1/250, exactly the expected activity. Thus, NO ANTIBODY WAS
INACTIVATED. We are at a loss to explain the direct titration result
since the A550/A280 was 0.22. We will repeat the titration, and also
try using Molecular Probes carboxytetramethylrhodamine succinimidyl
ester for our next try.

It turns out that the second antibody mentioned was entirely inactivated
during the purification process (before conjugation), presumably by
low pH elution. Thus, TRITC was unrelated to its problem.

-- 
/* - - - - - - - - - - - - - - - - - From - - - - - - - - - - - - - - - - - -
Eric Martz              emartz@titan.ucc.umass.edu    Professor of Immunology
                    Voice 413-545-2325, FAX 413-545-1578
    Morrill IVN Rm 203, Dept Microbiol, Univ Mass, Amherst MA 01003 USA
- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - */


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu