lysing solution

Arnold.Pizzey (a.pizzey@ucl.ac.uk)
Thu, 05 Sep 1996 13:22:20 +0000

Hi folks, try this,

Reagent A
60ul Formic acid
50 ml H2O

Reagent B
15.6mM Sodium Carbonate (anhydrous)
68.2mM Sodium Chloride
60.5mM Sodium Sulphate
(In H2O)

Reagent C
1% Paraformaldehyde in Phospate Buffered Saline
(although this could be in just about anything, including water)

Method:
To 50ul whole blood (I've tried this with EDTA,Citrate,Heparin), add:

300ul Reagent A - vortex for 10 Seconds,
then:
132.5ul Reagent b - vortex as above
then:
50ul Reagent C - Vortex as above

Samples may be stored on ice until analysis (note-no wash step)

I've used this technique on 10s of thousands of samples and
i) it gives v.good FS/SS plots (lymphs,monos,neuts can be easily seen)
ii) it works for the following:

CD11B/CD62/CD10/CD4/CD8/CD14/CD16 oh, and lots of others I can't
remember off the top of my head.
It also works with the DCF assay.

All the best
Arnold


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu