Re: pulse width

Thomas Delohery (t-delohery@ski.mskcc.org)
Wed, 31 Jul 1996 12:48:11 +0530

Alice Given brings up several good points when she wrote:

>Forward scatter, however, is even more complex. It's related to refractive
>index as well as to size of the particle. And, of more relevance, it is
>proportional to cross-sectional area of the particle (NOT to diameter). The
>larger the cross-sectional area of the particle, the more light passes through
>it and gets refracted around the obscuration bar to reach the photodector. So,
>even if refractive indices are constant from particle to particle and even if
>the particles are larger than the laser beam (both highly questionable), you
>would not expect a straight line relationship between FSC and pulse width, but
>rather a curve.

The forward light scatter for cells is a combination of refraction and
diffraction. On the other hand, forward scatter from plastic beads is
exclusively due to diffraction. Thats why the forward scatter signal will
always be greater for a cell than a bead of the same size. This doesn't
begin to address the complexities of obtaining quantitative "size" measurement
from scattered laser light.

my two bits,
td

be the particle in the beam.

--
==============================================================================
 Thomas Delohery                        | Internet: t-delohery@ski.mskcc.org
 Manager, Flow Cytometry Core Facility  |    Phone: (212) 639-8729
 Memorial Sloan-Kettering Cancer Center |      Fax: (212) 794-4019
 1275 York Ave. Box 98                  |
 New York, NY    10021                  |
==============================================================================
 


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu