> I have a user who has been assessing viability of neural cells using
> Molecular Probes Eukolight kit which uses Calcein AM for live cell
> measurement and Ethidium homodimer for dead cell measurement. My
> question is this: Since ethidium staining is related to leakage into
> the cell as a result of altered membrane integrity and calcein is
> related to cytoplasmic esterase activity, what are the kinetics with
> respect to the sequence of events? Is it possible for cells to stain
> with ethidium while still maintaining esterase activity; hence stain
> with both dyes, i.e. damaged but not dead? Since these samples were
> run on a FACscan which cannot compensateFL1 and FL3 and there is
> significant spectral overlap between these dyes it was not possible
> to look at the dual color staining by flow. However using
> fluorescence microscopy we definitely saw dual staining. Could it be
> that the cells were damaged but still had some esterase activity? I
> would appreciate any responses from people who have worked with this
> kit before.
I can't answer the actual question on esterase activity after cell
damage from first hand experience, but I can tell you how to do the
compensation so that you might answer it yourself on your FACScan:
If Ethidium fluorescence leaks into FL1 then it must leak more into
FL2. Set FL2-%FL3 to 0.0 to allow that leakage into FL2. You can
then remove Ethidium fluorescence from FL1 by adjusting FL1-%FL2.
Similarly, if Calcein fluorescence leaks into FL3, then it must leak
more into FL2. Set FL2-%FL1 to 0.0 to allow leakage into FL2. You can then
remove calcein fluorescence from FL3 by adjusting FL3-%FL2.
\ / < Flow Systems Laboratory
Frank Battye \__/ <<<<< The Walter & Eliza Hall Institute
----------!!<<<<<<<< Voice: 61_3_9345 2540
/!!\ <<<<< Fax: 61_3_9347 0852
o !! \ < IN:: "facs_copy@wehi.edu.au"
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web