Forward scatter, however, is even more complex. It's related to refractive
index as well as to size of the particle. And, of more relevance, it is
proportional to cross-sectional area of the particle (NOT to diameter). The
larger the cross-sectional area of the particle, the more light passes through
it and gets refracted around the obscuration bar to reach the photodector. So,
even if refractive indices are constant from particle to particle and even if
the particles are larger than the laser beam (both highly questionable), you
would not expect a straight line relationship between FSC and pulse width, but
rather a curve.
Let someone else take over from here!
Alice L. Givan
Englert Cell Analysis Laboratory
Dartmouth Medical School
Lebanon, New Hampshire
NH 03756 USA