Re: megakaryocytes
Howard Shapiro (hms@shapirolab.com)
Sat, 20 Jul 1996 21:18:35 -0400 (EDT)
>From about 1987 to the early 1990's, Dave Kuter, working with Bob Rosenberg
at MIT and using a Cytomutt with a 5 mW argon laser, developed and used a
megakaryocyte ploidy assay to aid in purifying platelet growth factor.
Cultured marrow cells were indirectly labeled with a polyclonal
anti-platelet antibody/fluorescein developing antibody and propidium iodide;
PI fluorescence was used as the trigger signal with the level set so that
diploid and tetraploid cells were ignored, and a rectangular acquisition
gate was set to include cells with high (8n or more) PI fluorescence and
high fluorescein fluorescence. Megakaryocytes accounted for approximately 1
in 1,000 nucleated cells in the samples, which were run at approximately
5,000 cells/sec (remember, most of the nucleated cells didn't trigger the
system). This allowed a histogram including 1,000 of the 8n to 64n
megakaryocytes to be accumulated in a few minutes.
(Kuter DJ, Greenberg SM, Rosenberg RD: Analysis of megakaryocyte ploidy in
rat bone marrow cultures. Blood 74:1952-62, 1989
Kuter DJ, Rosenberg RD: Regulation of megakaryocyte ploidy in vivo in the
rat. Blood 75:74-81, 1990
Kuter DJ, Beeler D, Rosenberg RD: The purification of megapoietin: a
physiological regulator of megakaryocyte growth and platelet production.
Proc Natl Acad Sci USA, 91:11104-8, 1994)
Assuming you had the nozzle hemodynamic problems solved, you could
implement a similar solution for viable sorting using Hoechst 33342 and
perhaps something more elegant in the way of antibodies.
-Howard
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web
http://www.cyto.purdue.edu
, EMAIL
cdrom3@flowcyt.cyto.purdue.edu
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web
http://www.cyto.purdue.edu
, EMAIL
cdrom3@flowcyt.cyto.purdue.edu