At 04:15 PM 7/9/96 -0400, you wrote:
>
>Simon Monard noted:
>> 50u mesh sounds a bit big for a 70u nozzle.
>
>Indeed, BDIS recommends using nozzles with a diameter that is four times
>larger than the diameter of the largest particle sorted! A 200 micron tip
>is available for the MacroSORT option (largest nozzle is 400 microns). For
>a 70 micron tip, I use a 20 micron mesh; admittedly not quite 4:1, but at
>least it gets out the visible clumps, pieces of tissue, etc. :)
>
>/\/\/\_ Eric Van Buren, vanburen%flovax.dnet@rocdec.roc.wayne.edu
>\ \ \ Karmanos Cancer Institute and Wayne State University,
> \_^_/ Immunology & Microbiology. Detroit, Michigan.
>
>
![]() |
![]() |
![]() |
![]() |
![]() |