Annette, Valarie, and flow list:
We attempted to ask this question using FITC-labelled antigens taken
up by normal mouse B cells in vitro. This was published in Mol. Immunol.
28:779 (1991). Basically we used the fact that chloroquine can neutralize
the low pH of endosomes, and developed a fixation technique with ethanol
the maintained all the internalized FITC-label. We found that in
solution, the fluorescence of our FITC-antigen was altered by a factor of
six, when pH was adjusted from pH 5.5 to pH 7.2. After internalization
(and stripping of surface bound material) chloroquine could only effect a
3 fold shift in fluorescence intensity. However, the same cells fixed in
ethanol and exposed to pH 5.5 and pH 7.2 had again the 6 fold shift.
This indicated to us that not all of the internalized material made it to
low pH compartments.
The work was done on a modified FACS II instument, but I'm sure the
FACScan is sensitive enough to do the same. This may be a useful technique
for your work, provided that you have a selective FITC marker that is
internalized to the compartment of interest. I'm not sure this
technique is capable of clear quantitative work, and I never pursued it
as such.
Denis Snider PhD
Associate Professor
McMaster University
--------------------------------------------------------
On Mon, 20 Nov 1995, Annette Arriens wrote:
> Hello all,
>
> A colleague who is not on the mailing list for this group wants to
> measure pH in endosomes and lysosomes using a FACScan flow
> cytometer. She's been looking for protocols without much success.
> Can anyone help? Information can be sent either to the list (for
> anyone else interested in the procedure) and/or to her directly to
> her at natale@ivi.ch . Her name is Valerie Natale. Cheers.
>
> Annette Arriens
>
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web