Re: Detection of Cy5

Dave Coder (dave@nucleus.immunol.washington.edu)
Fri, 3 Nov 95 16:23:45 -0800

Begin forwarded message:

Greetings from the U. of Vt. I've just received a call from an investigator
who is going to label platelets with a FITC conjugated marker and an
activation marker linked to Cy5. They are using a kit from Amersham LIFE
SCIENCE called FluoroLink-Ab Cy5 Labeling Kit.The detection of plts. is no
problem, but I've never worked with Cy5. Does anyone have experience with
this kit and/or running such specimens on the flow.

{No experience with the kit, but lots with Cy5. Although you haven't
specified how you're going to be using the reagents, I can provide some
generalities about Cy5. If it's a direct conjugate of the antibody, you need
to excite Cy5 with red light--either a HeNe laser (633nm) or a Kr laser
(647nm) or a diode laser emitting somewhere in this range. Cy5 fluorescence
is best viewed around 670nm; I believe that I had Omega make a 670BP20 for
about $150 some 5 years ago. FITC and Cy5 are a good pair with respect to
fluorescence because there is little appreciable emission overlap of the two
dyes.

{This means that you need a dual laser system or a mixed gas Ar/Kr laser
emitting 488nm and 647nm in order to view FITC and Cy5 associated with the
same cell.

{If you're making a direct conjugate of your own antibody with activated Cy5,
the conjugation is very easy to do. As I recall, separation of free dye from
dye/protein conjugate was very efficient over a G10 column; you may or may
not have to concentrate by dialysis. The conjugation was good in preserving
antibody activity. A dye/protein ratio of around 2 should avoid quenching
problems. Since excess dye will be washed away there will be no residual dye
to stick to tubing etc.

Dave Coder
dcoder@u.washington.edu


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu