> Related to the discussion on fluorescences beyond saturation, could anyone
> give me a vague idea of the typical concentration range over which directly
> labelled abs exhibit saturation? Or is this extremely variable depending on Ab
> type, cells, etc?
>
> TIA
>
It's essentially a function of antibody affinity.
[ab]+[ag]<>[abag]
For many of the monoclonal antibodies I work with saturation occurs at
about 5 to 10 ug/ml although some staturate at 1ug/ml and others don't
saturate until about 50ug/ml.
Of course we rarely work at equilibrium.
As an approximate guide the avidity of an antibody is the molar
concentration at which you see 50% saturation. More specifically it
should be the molar concentration of unbound antibody, but for most
antibodies free will be in excess over bound so free can be approximated
to input concentration.