>>>>Can someone please outline the physical attributes of PerCP, with respect
>>>>to that dyes requirements for excitation time, laser power requirements,
>>>>etc. How do these compare/contrast with the requirements of something like
>>>>FITC, PE, or APC?
>>>>For example, I see that PerCP works well when exited within the cuvette
>>>>of a FACScan, but not in a stream-in-air sorter, ostensibly due to the
>>>>longer time and lower laser-power requirements of the dye. Can someone give
>>>>me the real FACS? thanks.
PerCP is approximately 35Kd in size and has the following properties:
FITC PE PerCP APC
Excitation (max): 495nm 565nm* 470nm 650nm
Emission (max): 525nm 575nm 677nm 660nm
Optimal excitation: 80mW 80mW <15mW 460 mW**
Excitation time: ***see below
Quantum Yield: ?.65 ?0.94 No data ?.75
*PE has a soret band maxima at 470nm, which enables the 488 excitation.
**APC S/N at 460mW only twice the S/N obtained at 47mW when 633 nm excitation
was used (Loken et al, Cytometry 8:96-100 (1987).
***I don't have any specifics - but I think there may be some info in a chapter
written by Alan Waggoner, "Fluorescent Probes for Cytometry", Flow Cytometry and
Sorting, 2nd Ed. Melamen, Lindmo, and Mendelsohn Eds., Wiley Liss (1990) pp.
209-225.
Relative Spectral Overlap*:
FITC PE PerCP APC
FITC 1.000 0.200 0.003 0.000
PE 0.015 1.000 0.080 0.000
PerCP 0.000 0.000 1.000 0.05
APC 0.000 0.000 0.05 1.000
*Filter choices will of course change the relative amounts of compensation
required.
FYI: the molecule is extracted from a dinoflagellate. By contrast, PE and APC
are extracted from the phycobilisomes of either blue-green bacteria or various
rhodophytes.
Having said that PerCP is optimal on <15mW, if a 50mW laser on a stream in air
is used with appropriate filters, it is possible to use PerCP for highly
expressed antigen detection. NOTE: I wouldn't try and use HLA-DR PerCP on a
vantage!
I hope that this helps
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web