We also have noted problems of carryover with small (0.5 micron)
beads on an XL. They persist for many subsequent tubes when we
use the autoloader. I have found that following the bead tube
with 50% ethanol and then saline removes almost all of the
beads. This is a problem because it adds extra tubes, but
nothing else seems to work. The Coulter field technician was
completely surprised when I showed him this. Apparently their
usual tests are with larger beads (7-10 micron) and these do not
carryover. However, their definition of carryover is < 1.0 %.

Good luck, Tony Bakke (bakkea@ohsu.edu)


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu