Jacek Witkowski (jawit@amed01.amg.gda.pl)
Mon, 29 May 1995 18:45:59 +0200 (MET DST)
A caveat to all attempting to use Rhodamine R123 as G0/G1 marker in bone
marrow and lymphocytes:
Due to a presence of an universal membrane transporter P-glycoprotein, R123 is
actively eliminated by a subpopulation of human CD34+ hematopoietic
cells (cf. Chaudhary & Roninson, 1991, Cell, 66 : 85) and by some T cells
(in humans: Gupta et al., 1992, J. Clin. Immunol. 12: 451; in mice:
Witkowski & Miller, 1993, J. Imunol. 150: 1296). Our recent observations
(unpublished so far) indicate that also in human cord blood lymphocytes
the level of active R123 exclusion is quite high.
If one wants to use R123 as a mitochondrial marker (which it is!) in
these cells, in hope to see the increase in staining due to G0->G1 shift,
should bear the above in mind and use some kind of blocker of
P-glycoprotein activity (50 uM verapamil will do).
By the way, Hoechst 33342 is also pumped by cells expressing
P-glycoprotein activity (cf. Krishan 1987, Cytometry 8:642; Morgan et al,
1989, Br. J. Cancer 60: 339).
Good luck!
Jacek M. Witkowski MD, PhD
Department of Histology
Flow Cytometry Laboratory
Medical University of Gdansk, Poland
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web
http://www.cyto.purdue.edu
, EMAIL
cdrom3@flowcyt.cyto.purdue.edu