RE: Texas Red Laser Excitation

Howard Shapiro (hms@shapirolab.com)
Fri, 3 May 1996 17:06:38 -0400 (EDT)

Mark Kukuruga asks if Texas Red is excitable at 633 nm, by an
air-cooled He-Ne. 633 nm, no; air-cooled He-Ne, yes. Bob Hoffman and
I put my 594 nm air-cooled He-Ne into an Ortho instrument at the 1985
Hilton Head SAC meeting, and detected Texas Red immunofluorescence.
This was a very ecumenical experiment; Coulter lent us the emission
filter. I know Coulter has demonstrated configurations of the Elite
with the 594 nm laser, and, while you can't (as far as I know) get a
25-50 mW 594 nm He-Ne laser, you can get a 5-10 mW 594 nm He-Ne laser.
It would work in a system with efficient optics like the Elite or, if
B-D felt like installing one, in the FACSscan, FACScalibur, etc.
-Howard


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu