Fri, 28 Apr 1995 17:11:46 -0700

Hi,

We've upgraded our FACStar+ with the pulse processing+7th parameter option
(Fl3-2 on the second laser). Unfortunately the 7th parameter upgrade does
not deliver 7 parameter sorting. While sorting the SSC is disabled since
the 7th detector signal is routed through the Side Scatter electronics and
we end up with 6 parameter sorting. What we want is to be able to sort on
parameters 1-7. We have been told by BD that the only piece of hardware
providing that option is the SEM (Sort Enhancement Module). I briefly
looked through the SEM guide and I am not sure that for the money asked
(~$10K CDN) this is the best option. It seems unnecessary to me to add the
new sort control unit replacing the FACStar+ sorting card#10 just to add 1
sorting parameter (and I really don't care about nonrectangular gating +
other "options"). Are there any other hardware solutions we might consider?
If not, are there any people out there working with the 7th parameter
upgrade (I call it a downgrade!)+SEM configuration? How does the SEM
affect the dead time? What about sort purity and recovery? Is it like it
says in BD brochure that the sorting purity is ~95% at more than 5,000
cells/sec (interesting what do they mean by more.., sounds very ambitious,
isn't it?) and that the SEM "saves time by providing a pure population with
one pass" (they don't give numbers this time). Any response would be
greatly appreciated.

Best regards

Visia
Flow Cytometry
Terry Fox Lab
Vancouver
Visia Dragowska


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu