I response to Eric Martz comments:
--------------------------------------------------------
>Live rate displayed Rate saved to file Events lost
>on monitor to disk file
>3,200/sec 294 91%
>420/sec 357 15%
>
>Since the rate saved is actually slower at the higher rate, there is
>evidently a penalty for higher rates above and beyond the dead time
>per recorded event. I think everyone should do the kind of check you
>did. The conclusion (for our setup) is that concentrations of cells
>giving rates above 500/sec are simply wasting cells!
-------------------------------------------------------
According to a poster by Langlois and van den Engh presented at ISAC XVI,
March 1993: Both BD and MoFlo maximum data storage for 3-parameter list
mode was approx. 1000 events/sec.. A max.of 4000 events/sec. was observed
for "two-parameter histograms".
This should not be confused with the time required in signal processing
for sort-decision making - which is much faster. This is why I reported
trigger rates.
tom d.
"Never, for the sake of peace and quiet deny your own experience
or convictions"
Dag Hammarskjold
-- ============================================================================== Thomas Delohery | Internet: t-delohery@ski.mskcc.org Manager, Flow Cytometry Core Facility | Phone: (212) 639-8729 Memorial Sloan-Kettering Cancer Center | Fax: (212) 794-4019 ==============================================================================
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web