> PROBLEM: Profile data does not appear to contain any equivalent to the FCS
> "$TOT" keyword; That means that all the analysis software at my disposal
> cannot analyse the files, because they appear to contain NO events!
You're right, Pro2FCS files do not provide $TOT (in any FCS form).
I've modified MFI to work fine with them, though. Here's an excerpt
from a beta-test announcement for MFI I made on 1/13/95. (Subsequently,
MFI 3.4I has been released including these features.)
[BEGIN QUOTE]
A peculiarity of this [Pro2FCS] file structure is the absence of $TOT. I
calculate the total number of events from the parameter data sizes and the
offsets. If anyone is interested, ask and I can make available the C code
I use. I do this for all files anyway since in some cases the number of
events actually recorded in the data block is less than that given in
$TOT. [A warning message is emitted when there's a discrepancy.] I have
not been able to obtain a Pro2FCS file containing TIME as a parameter so
MFI may not interpret the TIME range properly when it first encounters such
a file.
[END QUOTE]
I still have never obtained a Pro2FCS file with TIME. If someone has
such, please put a sample in /incoming at flowcyt.bio.umass.edu for
my and MFI's edification.
/*- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
Eric Martz, Professor of Immunology emartz@microbio.umass.edu
Dept Microbiology Voice: 413-545-2325 FAX: 413-545-1578
Morrill IVN 203, Box 35720, Univ Massachusetts, Amherst MA 01003-5720
- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -*/
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web