Re: Resolution of dim fluorescence

Dennis_Young@CIS.ucsd.edu
Tue, 21 Mar 1995 16:13:00 -0800

The scaling of the histograms will effect where they intersect (display constant
*PERCENTAGE*).
One assumes non-specific binding is equal for both (or more) reagents.
The intersection method assumes also that false-positives are balanced by
false-negatives on either side of the inflection. You will sometimes see two
populations after subtraction; above and below the original histogram's peak.

What is the precision of the staining and how many replicates do you prepare
per experiment?

Flow cytometry *IS* limited in sensitivity (to about 5,000 FITC per particle)
and may not be the appropriate assay system. Positive/negative for fluorescence
does *NOT* mean that the antigen is or is not there!

Do you have this:
Flow Cytometry Titration of Retroviral Expression Vectors: Comparison of Methods
for Analysis of Immunofluorescence Histograms Derived from Cells Expressing Low
Antigen Levels, Sladek and Jacobberger, Cytometry, 14:23-31 ,1993)?

*****************************************************************************
* Dennis J. Young Voice : (619) 543-3928 *
* Flow Cytometry Core Facility FAX : (619) 543-7487 *
* University of California, San Diego USA e-mail: djyoung@ucsd.edu *
*****************************************************************************


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu