Despite Rob Chervenak's comments ("are these people NUTS"), it is not
only possible but in fact can be desirable to have individuals operate
the sorters for themselves ("self-operators", as we term them). In our
facility at Stanford (managed by Dave Parks), we have quite a number of
people who operate the sorters themselves. We do not have frequent problems
of the machines being "screwed up" by these operators... (Of course, it does
rarely occur). The reason for self-operating being desirable is two-fold: (1)
it allows our operators to be used for other users on other machines, making
more efficient use of the resources, and (2) it makes the users understand
what they are doing, and hopefully thereby do their experiments
properly. It is just as difficult for FACS operators to understand the
nuances of everyone's science as it is for scientists to understand the
nuances of sorting/analysis.

There are two primary reasons that the self-operater sorting is so
successful here. (1) The machines are constantly being using by the
skilled operators for those people who don't which to self-operate, so
they are constantly fine-tuned and maintained at a very high level.
(2) Dave has put considerable effort to defining stable sort conditions
on each machine, that do not vary. It is almost recipe-like in the
approach. In fact, it takes me no more than 5 minutes from the time I
sit in front of the machine (having focussed and calibrated it with
beads), to set up the sort conditions and have it sorting. We use
exactly the same nozzle frequency, drop delay, and first drop distance
for every sort. Then we simply adjust the sheath pressure so that the
inter-drop distances are a preset value. Voila: perfect sorts. I
routinely get 99.7% purity with >98% recovery. (Also, it usually takes
me no more than 5 minutes to start up and calibrate the machine).

Admittedly, I am a seasoned operator. But we have "beginners" in our
lab and other labs at Stanford, who can barely change the sheath fluid
reservoir, but can set up good sorting conditions in 10-15 minutes. We
have dozens of graduate students and post docs at Stanford who do their
own sorting, and quite a few more who do their own operating for
analysis. (And, they do this not only on a FacstarPlus, but also on a
FACS II!--not the most user-friendly of machines).

This facility is proof that self-operating can work extremely well.
The only requirements are a well-maintained facility and reasonable training:
something that you should have anyway if you employ a full-time operator!

mr


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu