We have looked at EMA staining (with UV activation) for marking
non-viable cells in samples which were subsequently fixed (as per
Reidy et al Cytometry 12, 133-9 '91). We found, as did Mark, a
background fluorescence on initially-viable cells that increased over
time roughly (where we have done a *very* approximate calibration
using Quantum beads) as follows:
FITC PE
DAY 0 <0.4 11. (All in thousands of molecules)
DAY 1 1.4 98.
DAY 7 5.5 122.
We did *not* see changes in light scatter as a result of EMA staining;
our fixation procedure reduces FSC by about 3%. Correlation between
EMA and PI for viability seemed OK.
An alternative stain we tried was 7-AAD. We intended to use the
protocol of Fetterhoff et al.(ISAC XVI '93 abstract 204B) where
Actinomycin-D is used after fixation to saturate DNA sites not already
binding 7-AAD, but found, surprisingly, that when using 7-AAD alone,
the problem of staining of the fixed but initially-viable and
therefore initially unstained cells did not arise. *No* increase in
fluorescence background of the initially-viable cells was seen over 7
days however the 7-AAD intensity dropped off by a decade or so.
\ / < Flow Systems Laboratory
Frank Battye \__/ <<<<< The Walter & Eliza Hall Institute
Robyn Muir ----------!!<<<<<<<< Voice: 61_3_345 2540
Dora Constantinou /!!\ <<<<< Fax: 61_3_347 0852
o !! \ < IN:: "facs_copy@wehi.edu.au"
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web